Mark Cannell sent the following addition to my explanation
but so far as I can see just to me, not the list. So I'll
add it here:
I said:
Oxygen is naturally a triplet molecule. Triplet-triplet reactions are particularly
likely to occur, and so a triplet excited state is more likely to get oxidised.
Mark adds:
Your answer is very close. The final part is that the triplet-triplet interaction produces two singlet molecules and singlet oxygen is very reactive. Since it is in close proximity to the fluorochrome it is likely to oxidize it.
As to bizarre explanations, I can't see the point. How do they
help? Singlet just means all electrons are paired with ones of
opposite 'spin' - think of them as tiny magnets, each one that
has its north pole upwards is matched with one with its south pole
upwards, so there's no overall magnetic field. Triplet means that
there are unpaired electrons. Mark has explained the reason for
the names, but they're not so important.
One possible decay route for an excited singlet molecule is to
go into a triplet state. It's relatively unlikely for an electron
to change its spin so usually not many will decay that way. But
since we're exciting a lot of molecules some will chance to take
that path. Again, getting from there back to a non-excited state
means changing an electron's spin, and since this is an infrequent
occurrence the molecule will remain in a triplet state for a while.
If nothing happens in this time the molecule will go back to the
ground state, unbleached, but it is very vulnerable to oxygen attack
while waiting in the triplet state.
None of this is difficult to follow. The explanation that 'It's easy
to switch to the dark side' is simply wrong. It's quite hard to
switch to the dark side, it's just that if you keep on exciting a
molecule, over and over, eventually it'll end up there.
The best analogy I can think of is a fly-paper. It's quite unlikely
that a fly will bump into the fly-paper, but when it does, it can't get
off. So eventually you'll get rid of all the flies in the room.
Guy
Optical Imaging Techniques in Cell Biology
by Guy Cox CRC Press / Taylor & Francis
http://www.guycox.com/optical.htm______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176 Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Tobias Baskin
Sent: Monday, 16 February 2009 11:47 PM
To:
[hidden email]
Subject: Re: Photobleaching mechanism question
John,
From the non physicist's point of view the answer could go something like this. If you have a power surge it can fry your computer. But if your computer is not plugged into the mains then it would take a very big power surge indeed to do the damage. A molecule in the ground state can of course be damaged by free radical attack but no more or less than other molecules. But once a molecule has absorbed a photon then it is not in the ground state any more. To continue my hoaky analogy, a chromophore in light is like your computer plugged in to the mains.
Hope this helps. The physicists (and musicians) can go for the triplets.
Tobias
>Hi Everyone, this question follows on from a helpful discussion that
>we had about photobleaching back in November. I have recently tried to
>explain to a group of colleagues about the mechanism of photobleaching.
>The answer is based on the transition of molecules from the excited
>singlet state (S1) to the triplet state (T1) which is long-lived and
>therefore more susceptible to bleaching by free radicals (my entire
>discussion of this is below).
>
>My question that arises from my attempted answer is: why are excited
>molecules more susceptible to oxidative attack than ground state
>molecules. I hope I'm not completely mucking up the mechanism here.
>Would the physicists out there please help.
>
>Thanks, John.
>
>The original answer: When excited, fluorophores generally transition
>from singlet ground state (S0) to singlet excited state (S1).
>Relaxation from S1 to S0 results in emission of heat and light
>(fluorescence). Lifetime in S1 is in the nano to pico second range and
>allows very little time for the excited molecule to interact with free
>radicals. Periodically, however, an excited molecule will do a
>transition from S1 to the triplet excited state (T1 - the physics of
>this is a bit difficult to understand). T1 is a very long-lived state -
>molecules can remain in T1 for up to the microsecond range - i.e. a
>thousand to a million times longer than for normal S1 state. It is
>during this long T1 state that molecules are attacked by free radicals
>and destroyed.
>
>--
>Runions signature
>
>(Sent from my cra%#y non-Blackberry electronic device that still has
>wires)
>
>
>
>*********************************
>John Runions, Ph.D.
>School of Life Sciences
>Oxford Brookes University
>Oxford, UK
>OX3 0BP
>
>email: <mailto:
[hidden email]>
[hidden email]
>phone: +44 (0) 1865 483 964
>
><
http://www.brookes.ac.uk/lifesci/runions/HTMLpages/index.html%21>Runions'
>lab web site
>
>
>
>Visit <
http://www.illuminatedcell.com/ER.html>The Illuminated Plant
>Cell dot com Oxford Brookes Master's in
><
http://www.brookes.ac.uk/studying/courses/postgraduate/2007/bmt>Bioima
>ging
>with Molecular Technology
No virus found in this incoming message.
Checked by AVG.
Version: 7.5.552 / Virus Database: 270.10.24/1954 - Release Date: 15/02/2009 6:09 PM
No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.552 / Virus Database: 270.10.24/1954 - Release Date: 15/02/2009 6:09 PM