Hi all,
We have all seen ray diagrams like.
http://www.antonine-education.co.uk/New_items/MUS/images/ray_diag_6.gif
Light from a point on the axis and in the focal plane diverges
until some of it hits the (perfect) lens and emerges as a parallel ray
bundle on the other side. However, things become a little more complex
when the focal-plane emitter is moved off the axis, as in
http://www.phg.ulg.ac.be/Hololab/fichiers/Graphic1.jpg
Now the emergent parallel ray bundle is at an angle to the axis.
So you would think that, as long at the next element in the optical
system (the tube lens) was sufficiently large enough (or close enough)
to collect the angled ray bundle created by light emerging from the
focus plane at the edge of the field of view, then all the light from
the field of view would be focused into the Intermediate Image. Then
it would pass through the eyepiece/camera-coupler and end up in the
recorded image.
But what about light emerging from above or below the focal
plane? Although I couldn't find an on-line sketch, it is clear that
instead of producing a parallel emerging ray bundle, it will produce
one that either diverges or converges (resp). So some of these rays,
particularly those from points both off axis and out of focus, will
leave the objective at angles even larger than those from the edge of
the field of view at the plane of focus. How much more will depend on
how far off axis and away from the focal plane they are, but when one
focuses tens of micrometers into a thick, fluorescent specimen, it can
be quite a large angle.
As we want to place reasonable limits on the diameter of the
black metal tube between the objective and the tube lens, (and on the
diameter of the tube lens itself), then compromises must be made and
some of the light from out-of-focus, off-axis sources will strike the
wall and be lost. The general term for this loss is
"Vignetting".
I know nothing about the actual changes that Olympus has
apparently incorporated in their LV200 except what I found at
http://www.olympus-europa.com/corporate/1696_1948.htm
However, as I remember it, bioluminescence has a very low signal
and the specimens are often quite 3-dimensional. To the extent that
this is true and that a high NA objective is used to collect the
signal, much of this signal may be out of focus. So I guess that
having a tube lens with a large diameter will collect more of this
light . Of course, it won't focus this extra light very well (the out
of focus sources will still look like pale, diffuse blobs. You can't
get around this.) but the total number of photons in the image will be
larger with this large tube lens than with a "normal"
one.
Two last notes:
1) If the tube lens has a larger diameter, the cone of
in-focus-light that converges from it to focus at the image plane will
have a larger half angle, and hence a higher NA. However, unless the
magnification of the entire system is very low indeed (i.e. using a
1-3x infinity objective), then this tube NA has no affect on the
sharpness of the final image (unless the large NA tube lens is
not properly corrected to produce near-diffraction-limited
performance at this NA: in which case the final image may be less
sharp).
2) To some extent, all microscopes have this limitation
(off-axis, out-of-focus light is preferentially lost) but it will be
particularly severe when using high-NA objectives of lower
magnification because these have larger fields of view. Indeed,
vignetting and other limitations on the performance of objectives are
discussed (with better figures!) in Chapters 7 and 11 of the Handbook
of Biological Confocal Microscopy.
Cheers,
Jim Pawley
--
**********************************************
Prof. James B.
Pawley,
Ph. 608-263-3147
Room 223, Zoology Research
Building,
FAX 608-265-5315
1117 Johnson Ave., Madison, WI, 53706
[hidden email]
3D Microscopy of Living Cells Course, June 13-25, 2009, UBC,
Vancouver Canada
Info: http://www.3dcourse.ubc.ca/
Applications due by March 15,
2009
"If it ain't
diffraction, it must be statistics." Anon.