Hello:
Don't mix Zeiss Immersol 518F and Cargille Type 37 immersion oils (the SOP I
wrote was clear on this!). I tried cleaning it but I had no luck. If I put a
drop of Zeiss oil on that objective and put a coverslip on, I can clearly see a
region which does not mix when I image the oil area/volume under a different
microscope!
I clean all objectives between batches of Zeiss oil just in case.
Bye!