Re: Overnight scan of human fibroblasts

Posted by Armstrong, Brian on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Overnight-scan-of-human-fibroblasts-tp2543999p2545052.html

Zoltan, in my experience the 2P/Ti:Sa lasers are still quite hard on the cell that you are imaging. I think the real benefit of 2P in live-cell is that you do not bleach the cells that are out of the focal spot of the 2P laser.

 

How to help you now: image less often, the total amount of energy you put into the sample can be greatly reduced by acquiring less often (I’m sure you must have thought of this already). Another idea is to use dyes that are Far-red (such as Draq5?) instead of the dyes that excite at UV or at V (405nm). I assume you are imaging nuclear staining of some kind. UV wavelengths are always going to damage your tissues more than the Red/IR wavelengths used to excite dyes in the far red range.

Cheers,

 

Brian D Armstrong PhD

Light Microscopy Core Manager

Beckman Research Institute

City of Hope

Dept of Neuroscience

1450 E Duarte Rd

Duarte, CA 91010

626-256-4673 x62872

http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Leigh Silvester
Sent: Friday, March 27, 2009 6:07 AM
To: [hidden email]
Subject: Re: Overnight scan of human fibroblasts

 

One answer, although not helpful in the short run is to use a tuneable infra red laser with multiphoton capability.

This can actually excite a wide range of fluorochromes. The excitation wavelength is approximately half that of the set wavelength.

Hence an IR laser pulsing nicely at 750 ish nm will excite DAPI beautifully.

 

As a result of the wavelength you don't get any DNA damage - just slightly warmed cells.

 

However, these lasers do not come cheap, and they are hard work/expensive to keep in running condition - although newer ones may be better.

 

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Zoltan Cseresnyes
Sent: 27 March 2009 12:19
To: [hidden email]
Subject: Overnight scan of human fibroblasts

Dear All,

One of our users would like to scan cultured human fibroblasts (focusing only on the nucleus) for about 15-17 hours at a time, 6 um Z stack (7 steps), 3-5 min time resolution, in 5% CO2 at 37C.  We have done a few very successful experiments this way already, using our SP5 inverted with an environmental chamber (The Box).  We used 458 nm or 488 nm excitation in those successful experiments at low laser power (<=20%).  Last night we tried 405 nm excitation (20% laser power) as well because we were attempting to do a live FRET scan; the scanned cells all died after about 2 hours.  Do you have any suggestions about how to protect these cells from photodamage for 15-17 hours?  We are trying to keep our lasers low, we were doing 20% power on both the 405 and 514 nm lines.  The temperature is fairly stable at 37C (+-0.8C perhaps) and the CO2 is okay as judged by the colour of the phenol red and the excellent survival of the cells in the non-scanned regions.    In one experiment last week we used 458 nm at 48% power (we had a poor fluorescence signal that time but we tried anyway) and then the scanned cells also died after about 6 hours.  These findings seem to point towards photodamage, hence my enquiry to this excellent forum of experts!
  Thanks very much,

Zoltan

--

Zoltan Cseresnyes
Facility manager, Imaging Suite
University of Cambridge, UK


---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.

---------------------------------------------------------------------