Re: microbead calibration to set
psf?
Hi all,
I think that Eric is right. Most specimens do have suitable small
features inside of them. Indeed, standardless, bootstrap (blind)
deconvolution depends on this in a way fact (although it can also
extract PSF info from the images of larger features as long as they
have sharp edges).
I would just like to add that 3D confocal images made by
collecting the light that is backscattered by small refractive index
anomalies are a particularly good source of point-like features.
If you have never collected a BSL image, all that is necessary is
to use some sort of beam splitter that will pass at least half of the
light at laser wavelength and remove the barrier filter. It is true
that this may reveal a signal largely consisting of light reflected
from the various glass-water and glass air interfaces in your optical
system but the confocal pinhole will eliminate most of these if they
are not located near image planes. The most annoying of these
reflections will be that from the glass-water interface on which the
cell may be resting. So focus a few microns into the cell.
There you will find a bewildering array of small scattering
features, the images of which should be suitable for determining the
excitation PSF. And all for free. No extra exposures needed, as it is
quite possible to collect the BSL and fluorescent signals
simultaneously and in a non-interfering manner
Such as optimized collection system is diagrammed in Figure 2.7
of the Handbook.
Cheers,
Jim Pawley
**********************************************
Prof. James B.
Pawley,
Ph. 608-263-3147
Room 223, Zoology Research
Building,
FAX 608-265-5315
1117 Johnson Ave., Madison, WI, 53706
[hidden email]
3D Microscopy of Living Cells Course, June 13-25, 2009, UBC,
Vancouver Canada
Info: http://www.3dcourse.ubc.ca
Applications still being accepted as space is
available.
"If it ain't diffraction, it must be statistics."
Anon.
Sorry for the self promo but on this
subject maybe some of you could be interested in the following
appproach?
It is indeed difficult to get proper
beads but also to get them into place if one considers that, for deep
imaging, the specimen is also part of the optical system!
Cheers
Eric
1: Biophys J. 2003
Dec;85(6):3991-4001.
Image-adaptive deconvolution for
three-dimensional deep biological imaging.
de Monvel JB, Scarfone E, Le Calvez S, Ulfendahl M.
Center for Hearing and Communication Research, Karolinska Institutet,
Stockholm,
Sweden. [hidden email]
Deconvolution algorithms are widely used in conventional fluorescence
microscopy,
but they remain difficult to apply to deep imaging systems such as
confocal and
two-photon microscopy, due to the practical difficulty of measuring
the system's
point spread function (PSF), especially in biological experiments.
Since a
separate PSF measurement performed under the design optical conditions
of the
microscope cannot reproduce the true experimental conditions
prevailing in situ,
the most natural approach to solve the problem is to extract the PSF
from the
images themselves. We investigate here the approach of cropping an
approximate
PSF directly from the images, by exploiting the presence of small
structures
within the samples under study. This approach turns out to be
practical in many
cases, allowing significantly better restorations than with a design
PSF obtained
by imaging fluorescent beads in gel. We demonstrate the advantages of
this
approach with a number of deconvolution experiments performed both
on
artificially blurred and noisy test images, and on real confocal
images taken
within an in vitro preparation of the mouse hearing organ.
Eric Scarfone, PhD, CNRS,
Center for Hearing and communication Research
Department of Clinical Neuroscience
Karolinska Institutet
Postal Address:
CFH, M1:02
Karolinska Hospital,
SE-171 76 Stockholm, Sweden
Work: +46 (0)8-517 79343,
Cell: +46 (0)70 888 2352
Fax: +46 (0)8-301876
email: [hidden email]
http://www.ki.se/cfh/
----- Original Message -----
From: Guy Cox <[hidden email]>
Date: Tuesday, March 31, 2009 4:08 pm
Subject: Re: microbead calibration to set psf?
To: [hidden email]
> Europium chelates are phosphorescent rather than luminescent
-
> that is to say, they have a very long - millisecond -
fluorescence
> lifetime. That should still work in widefield but would be
> hopeless for confocal. Pretty much unfadeable, which is a
plus.
>
> Tetraspeck bead are for calibrating colocalization rather than
PSF
> - having multiple fluorochromes on one bead must reduce the
> intensity at any one wavelength. As I said before, I've used
bead
> from Bangs Labs and I've had good images from beads down to
60nm.
> (Though at that scale it requires care and patience).
>
>
> Guy
>
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox CRC Press / Taylor & Francis
> HYPERLINK
>
"http://www.guycox.com/optical.htm"http://www.guycox.com/optical.htm______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176 Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
> HYPERLINK
"http://www.guycox.net/"http://www.guycox.net
>
>
>
> _____
>
> From: Confocal Microscopy List
> [mailto:[hidden email]] On Behalf Of David
Knecht
> Sent: Tuesday, 31 March 2009 11:57 PM
> To: [hidden email]
> Subject: Re: microbead calibration to set psf?
>
>
> I am looking for some very bright 100nm beads. I tried to
look
> for the beads you mention and it looks like Duke Scientific
no
> longer exists and is now part of hte Thermo conglomerate. The
> closest thing I found to what you describe is the Europium
chelate
> 100nm beads. Is that what you were referrring to? THey
provide
> very little data on fluorescence properties and they are
unusual
> in apparently being broad excitation spectrum but there is no
> emission spectrum in their literature I can find. Since they
> fluoresce maximally at 613nm it would seem they would be a
poor
> fit for most FITC, TRITC, CY3 or CY5 type filter sets. Can
you
> clarify your experience with them? Dave
>
> On Mar 31, 2009, at 3:36 AM, Mariette P.C. Kemner - van de
Corput
> wrote:
>
> I have good experience with fluorescent beads from Duke
> scientific. The
> 100nm beads are very bright and are perfect for determining
the
> PSF. I use
> the red, green and blue 100nm beads to determine the PSF in
three
> channelsand to deconvolved my 3 colour FISH images.
> I have trouble finding the 100nm and 200nm TetraSpeck beads on
my
> slidebeacuse they are rather weak in intensity. So that's why
I
> switched to the
> single coloured 100nm beads from DS. The 500nm TetraSpeck
beads
> are very
> good and I use them to determine the chromatic shift in 4
> channels. Once
> made, bead slides can be used for a long period of time.
>
> Mariette
>
> ____________________________________________
>
> Dr. M.P.C. Kemner-van de Corput,
> ____________________________________________
>
> MGC - Dept. of Cell Biology & Genetics
> Erasmus Medical Center
> Dr. Molewaterplein 50, 3015 GE Rotterdam
> POB 2040, 3000 CA Rotterdam, The Netherlands
>
> Office: H-Ee751; tel: +31 10 704.3949
> Lab: H-Ee710; tel lab: +31 10 704.3315
> tel secr: +31 10 704.3169
> ____________________________________________
>
> HYPERLINK
>
"http://www2.eur.nl/fgg/ch1/cellbiology/"http://www2.eur.nl/fgg/ch1/cellbiology/http://www.thesis.kemner.biz/
> ____________________________________________
>
> Op Di, 31 maart, 2009 1:59 am, schreef Matiar Jafari:
>
>
> hey again,
>
>
>
> I'm trying to find out how and what the best way to setup my own
psf
>
>
> is in AutoQuant. I know i need microbeads and im wondering what
brand
>
>
> and what product i need and then how i need to shoot it and load
it.
>
>
> Also what diameter does it need to be. A cat# would be great too
also
>
>
> i don't know if it matters on what im shooting but just to put it
out
>
>
> there im shooting gfp slices. Sorry if this is vague. Any help
would
>
>
> be greatly appreciated. Thanks in advance
>
>
>
> Thank You
>
>
> --
>
>
> Matiar Jafari
>
>
>
>
> Dr. David Knecht
> Department of Molecular and Cell Biology
> Co-head Flow Cytometry and Confocal Microscopy Facility
> U-3125
> 91 N. Eagleville Rd.
> University of Connecticut
> Storrs, CT 06269
> 860-486-2200
> 860-486-4331 (fax)
>
>
>
>
> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.33/2031 - Release
Date:
> 30/03/2009 5:56 PM
>
>
>
> No virus found in this outgoing message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.33/2031 - Release
Date:
> 30/03/2009 5:56 PM
>
>
--