I think
Zeiss’s comment that repeated use of ethanol will damage Zeiss’s
lens cement was just that – use it a few times a day and the Zeiss
objective will probably fail in 6 months or so [and this might be the case for
many solvents, ethanol is simply one more readily at hand*]. Use ethanol every
month or so and chances are the objective will last a lot longer [and fail for
another reason]. Use ethanol on a very elderly microscope where the lenses are
mounted in say gum resin though and you will destroy the objective pretty much
instantly [or at least get a nasty smear of gum resin all over the clear bit] -
hence some people’s historical aversion is justified. I did mention to a
Zeiss rep why did he just use 70% ethanol on our new 100x zillion NA TIRF
objective when Zeiss say that repeated use will damage the objective, he said,
well repeated use will damage the objective, but once or twice won’t
matter [and then I thought ‘well I suppose it’s not actually his
£8,000 objective’]. Other manufacturer’s actually recommend
ethanol, e.g. Olympus. However immersion oil doesn’t dissolve in ethanol
that well, hence another reason for the recommended use of other solvents, e.g.
Petroleum ether - see http://instrument-support.nikonusa.com/app/answers/detail/a_id/10379
*from the micro/primer site:
"In the past, solvents have been routinely employed for
nearly any cleaning task in microscopy, and particularly for removal of
immersion oil. Potential problems associated with solvent cleaning are
sufficiently serious that the best current approach in cleaning the microscope
is to use solvents only when absolutely necessary, essentially as a last resort
rather than a first step. The issue of the use of solvents is
further complicated and confused by contradictory recommendations in the
scientific literature, as well as by differences in manufacturers' technical
publications. Although alcohol and xylene are widely recommended as lens
cleaning solvents, they are also named as being harmful to both the mechanical
and optical components of many microscopes. Because of the variation in solvent
recommendations, and the likelihood that some of the materials used in the instrument
components are not known to the user, it is prudent to restrict use of any
solvent to an absolute minimum.”
i.e. have a look at: http://micro.magnet.fsu.edu/primer/anatomy/cleaning.html
Most of this discussion of solvents
only applies to immersion oil objectives, where the oil isn’t miscible in
water. Our inverted microscopes with all air objectives are rarely cleaned,
otherwise it’s just a blow with a puffer. On microscopes where oil and
air co-exist it’s often immersion oil contamination of the air objective
that’s the problem, so it’s solvents again. In the days when only
my group operated the microscopes, with all our oil immersion free upright
microscopes the objectives almost never needed cleaning.
For other spillages such as culture medium
[inverted microscope again] some solvents will probably fix biological muck
onto the lens, and there’s the salts content, so the use of water based
cleaners has been suggested [e.g. even breathing on the lens and then lens
tissue, using optical/glass cleaning solutions]. Water drying onto the lens is
a disaster though. Some even recommend things like breaking polystyrene foam
[to get a clean surface] and gently rubbing the [oil free] lens with that. Or
there’s Sparkle - whatever that is, here in the UK it was a silicon based
furniture polish [yuk] not a commercial window glass cleaner. That’s the
problem with industrial cleaners, who knows what’s in them or whether the
constituents have been modified – you could try it on an inconspicuous
area of the objective lens first, I suppose. Presumably optical lens cleaners
are glass friendly though, and many use glass cleaning products with no
reported problems. All the links in the previous posts [below] give loads of
ideas for cleaning objectives [when necessary].
http://books.google.com/books?id=Dhn2KispfdQC&pg=PA51&lpg=PA51&dq=petroleum+spirit+cleaning+objectives&source=bl&ots=JxlHfCVuF5&sig=NTJt3Ol66sgsB8gluF1eKpeXLCc&hl=en&ei=DfvRSdGpI5WrjAflkvjlBg&sa=X&oi=book_result&resnum=1&ct=result
http://www.microscopy-uk.org.uk/mag/indexmag.html?http://www.microscopy-uk.org.uk/mag/artfeb04/cdclean.html
www.olympus.co.uk/microscopy/images/illum_cleaning.pdf
http://www.olympusmicro.com/primer/techniques/fluorescence/troubleshoot.html
Keith
---------------------------------------------------------------------------
Dr Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory 00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt Drive,
Oxford OX3 7BN,
United Kingdom.
Telephone: +44 (0)1865 287568
Email: [hidden email]
Web-pages: http://www.well.ox.ac.uk/cytogenetics/
From:
Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Chris Tully
Sent: 30 March 2009 16:40
To:
[hidden email]
Subject: Re: Cleaning lens.
Dear all,
While working for a Leica Microsystems dealer the local field service engineer
(factor trained) used a sequence of ethanol and heptane to clean truly dirty
lenses. For standard cleaning a lens wipe and Sparkle was his
recommendation. But for dried oil or the like he would graduate to cotton
swabs and either ethanol then heptane or a 50:50 mix of the two.
Chris
Chris Tully
Microscopy and Image Analysis Expert
[hidden email]
240-888-1021
http://www.linkedin.com/in/christully
On Mon, Mar 30, 2009 at 10:12 AM, Ian Montgomery <[hidden email]>
wrote:
Keith,
Methylated spirit that’s what he said, although I still prefer and use
ether when necessary.
Ian.
Dr. Ian Montgomery,
Histotechnology,
I.B.L.S. Support Unit,
Thomson Building,
University of Glasgow,
Glasgow,
G12 8QQ.
From:
Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Keith Morris
Sent: 30 March 2009 14:12
To: [hidden email]
Subject:
Re: Cleaning lens.
Are you sure the Zeiss Engineer didn’t say
‘petroleum spirit’? Methylated spirit is mainly ethanol, and so
best avoided - the Axiovert 100 manual says repeated use of 70% ethanol will
damage the objectives [but you can use it if you want]. Generally the faster
the solvent evaporation from the lens/cement area the better, hence the
suggestion of the solvent [pure] diethyl ether by many [and that’s what I
use].
‘Zeiss cleaning mixture L’, which the
engineer’s now use since diethyl ether has been withdrawn from their kit,
is 90% by volume ‘benzoline’ [petroleum ether sometimes called
medical alcohol] and 10% ‘isopropanol’ [2-proponal, dimethyl
carbinol, 2-hydroxyproparne]. The bottle says ‘Clean the optics by moving
in circles, slight pressure should be exerted on optics during cleaning’.
Petroleum ether or spirit isn’t the same as the diethyl ether
solvent/anaesthetic often used to clean objectives, but apparently it does the
job for Zeiss optics.
Keith
---------------------------------------------------------------------------
Dr Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory 00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt Drive,
Oxford OX3 7BN,
United Kingdom.
Telephone: +44 (0)1865 287568
Email: [hidden email]
Web-pages: http://www.well.ox.ac.uk/cytogenetics/
From:
Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Ian Montgomery
Sent: 30 March 2009 12:28
To: [hidden email]
Subject: Cleaning lens.
In one of our teaching labs many years ago a student complained they were
having a problem with the x100 OI objective and sure enough the image was lousy.
I cleaned the objective and slide then re-applied a spot of oil and still the
image was lousy. I then asked the student how exactly they had set up the
microscope. Shock horror, my world collapsed. They had unscrewed the objective,
filled it with oil, screwed it back on then put a drop on the slide. After
weeks of trying to clean the objective it went into the trash as beyond
economic repair.
Cleaning objectives, I use the fluid recommended by the local Zeiss engineer,
90% methylated spirit and 10% isopropanol.
Ian.