I think
Zeiss’s comment that repeated use of ethanol will damage Zeiss’s lens cement was
just that – use it a few times a day and the Zeiss objective will probably fail
in 6 months or so [and this might be the case for many solvents, ethanol is
simply one more readily at hand*]. Use ethanol every month or so and chances are
the objective will last a lot longer [and fail for another reason]. Use ethanol
on a very elderly microscope where the lenses are mounted in say gum resin
though and you will destroy the objective pretty much instantly [or at least get
a nasty smear of gum resin all over the clear bit] - hence some people’s
historical aversion is justified. I did mention to a Zeiss rep why did he just
use 70% ethanol on our new 100x zillion NA TIRF objective when Zeiss say that
repeated use will damage the objective, he said, well repeated use will damage
the objective, but once or twice won’t matter [and then I thought ‘well I
suppose it’s not actually his £8,000 objective’]. Other manufacturer’s actually
recommend ethanol, e.g. Olympus. However immersion oil doesn’t dissolve in
ethanol that well, hence another reason for the recommended use of other
solvents, e.g. Petroleum ether - see http://instrument-support.nikonusa.com/app/answers/detail/a_id/10379
*from the micro/primer site:
"In the past, solvents have been
routinely employed for nearly any cleaning task in microscopy, and particularly
for removal of immersion oil. Potential problems associated with solvent
cleaning are sufficiently serious that the best current approach in cleaning the
microscope is to use solvents only when absolutely necessary, essentially as a
last resort rather than a first step. The issue of the use of solvents is
further complicated and confused by contradictory recommendations in the
scientific literature, as well as by differences in manufacturers' technical
publications. Although alcohol and xylene are
widely recommended as lens cleaning solvents, they are also named as being
harmful to both the mechanical and optical components of many microscopes.
Because of the variation in solvent recommendations, and the likelihood that
some of the materials used in the instrument components are not known to the
user, it is prudent to restrict use of any solvent to an absolute
minimum.”
i.e. have a look at: http://micro.magnet.fsu.edu/primer/anatomy/cleaning.html
Most of this discussion of solvents only
applies to immersion oil objectives, where the oil isn’t miscible in water. Our
inverted microscopes with all air objectives are rarely cleaned, otherwise it’s
just a blow with a puffer. On microscopes where oil and air co-exist it’s often
immersion oil contamination of the air objective that’s the problem, so it’s
solvents again. In the days when only my group operated the microscopes, with
all our oil immersion free upright microscopes the objectives almost never
needed cleaning.
For other spillages
such as culture medium [inverted microscope again] some solvents will probably
fix biological muck onto the lens, and there’s the salts content, so the use of
water based cleaners has been suggested [e.g. even breathing on the lens and
then lens tissue, using optical/glass cleaning solutions]. Water drying onto the
lens is a disaster though. Some even recommend things like breaking polystyrene
foam [to get a clean surface] and gently rubbing the [oil free] lens with that.
Or there’s Sparkle - whatever that is, here in the UK it was a silicon based
furniture polish [yuk] not a commercial window glass cleaner. That’s the problem
with industrial cleaners, who knows what’s in them or whether the constituents
have been modified – you could try it on an inconspicuous area of the objective
lens first, I suppose. Presumably optical lens cleaners are glass friendly
though, and many use glass cleaning products with no reported problems. All the
links in the previous posts [below] give loads of ideas for cleaning objectives
[when necessary].
http://books.google.com/books?id=Dhn2KispfdQC&pg=PA51&lpg=PA51&dq=petroleum+spirit+cleaning+objectives&source=bl&ots=JxlHfCVuF5&sig=NTJt3Ol66sgsB8gluF1eKpeXLCc&hl=en&ei=DfvRSdGpI5WrjAflkvjlBg&sa=X&oi=book_result&resnum=1&ct=result
http://www.microscopy-uk.org.uk/mag/indexmag.html?http://www.microscopy-uk.org.uk/mag/artfeb04/cdclean.html
www.olympus.co.uk/microscopy/images/illum_cleaning.pdf
http://www.olympusmicro.com/primer/techniques/fluorescence/troubleshoot.html
Keith
---------------------------------------------------------------------------
Dr
Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory
00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt
Drive,
Oxford OX3 7BN,
United Kingdom.
Telephone: +44
(0)1865 287568
Email: [hidden email]
Web-pages:
http://www.well.ox.ac.uk/cytogenetics/
From: Confocal Microscopy List
[mailto:[hidden email]] On
Behalf Of Chris Tully
Sent: 30 March 2009 16:40
To:
[hidden email]
Subject: Re: Cleaning
lens.
Dear all,
While working for a Leica
Microsystems dealer the local field service engineer (factor trained) used a
sequence of ethanol and heptane to clean truly dirty lenses. For standard
cleaning a lens wipe and Sparkle was his recommendation. But for dried oil
or the like he would graduate to cotton swabs and either ethanol then heptane or
a 50:50 mix of the two.
Chris
Chris Tully
Microscopy
and Image Analysis Expert
[hidden email]
240-888-1021
http://www.linkedin.com/in/christully
On Mon, Mar 30, 2009 at 10:12 AM, Ian Montgomery <[hidden email]>
wrote:
Keith,
Methylated spirit that’s what he said, although I still prefer and use ether
when necessary.
Ian.
Dr. Ian
Montgomery,
Histotechnology,
I.B.L.S. Support
Unit,
Thomson Building,
University of Glasgow,
Glasgow,
G12
8QQ.
From: Confocal Microscopy List
[mailto:[hidden email]] On Behalf Of Keith Morris
Sent: 30 March 2009
14:12
To: [hidden email]
Subject: Re: Cleaning
lens.
Are you sure the Zeiss
Engineer didn’t say ‘petroleum spirit’? Methylated spirit is mainly ethanol, and
so best avoided - the Axiovert 100 manual says repeated use of 70% ethanol will
damage the objectives [but you can use it if you want]. Generally the faster the
solvent evaporation from the lens/cement area the better, hence the suggestion
of the solvent [pure] diethyl ether by many [and that’s what I
use].
‘Zeiss cleaning mixture
L’, which the engineer’s now use since diethyl ether has been withdrawn from
their kit, is 90% by volume ‘benzoline’ [petroleum ether sometimes called
medical alcohol] and 10% ‘isopropanol’ [2-proponal, dimethyl carbinol,
2-hydroxyproparne]. The bottle says ‘Clean the optics by moving in circles,
slight pressure should be exerted on optics during cleaning’. Petroleum ether or
spirit isn’t the same as the diethyl ether solvent/anaesthetic often used to
clean objectives, but apparently it does the job for Zeiss optics.
Keith
---------------------------------------------------------------------------
Dr
Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory
00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt
Drive,
Oxford OX3 7BN,
United Kingdom.
Telephone: +44
(0)1865 287568
Email: [hidden email]
Web-pages: http://www.well.ox.ac.uk/cytogenetics/
From: Confocal Microscopy List
[mailto:[hidden email]] On Behalf Of Ian Montgomery
Sent: 30 March 2009 12:28
To: [hidden email]
Subject: Cleaning
lens.
In one of our teaching labs many years ago a student complained they were having
a problem with the x100 OI objective and sure enough the image was lousy. I
cleaned the objective and slide then re-applied a spot of oil and still the
image was lousy. I then asked the student how exactly they had set up the
microscope. Shock horror, my world collapsed. They had unscrewed the objective,
filled it with oil, screwed it back on then put a drop on the slide. After weeks
of trying to clean the objective it went into the trash as beyond economic
repair.
Cleaning objectives, I use the fluid recommended by the local Zeiss engineer,
90% methylated spirit and 10% isopropanol.
Ian.
Dr.
Ian Montgomery,
Histotechnology,
I.B.L.S. Support
Unit,
Thomson
Building,
University of
Glasgow,
Glasgow,
G12
8QQ.
No virus found in this incoming message.
Checked by
AVG.
Version: 7.5.557 / Virus Database: 270.11.35/2033 - Release Date:
31/03/2009 1:05 PM
No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.35/2033 - Release Date: 31/03/2009 1:05 PM
| Free forum by Nabble | Edit this page |