You won't be able to separate GFP and YFP reliably without a spectral
system; the emission wavelengths aren't separated by enough to do this
with conventional filters.
There are other fluorescent protein combinations you could consider
which might be better behaved, such as one of the new BFPs or a
UV-excited GFP like Sapphire which would let you do 4 proteins with
better separation between channels. I think you can also multiplex CFP,
GFP, mkOrange, and a far red FP like TagFP635 or tHcRed. You could
probably add BFP or Sapphire to that mix to do 5 proteins, although you
will probably start getting some crosstalk you need to deal with.
Kurt
Christian wrote:
>
> Recently a new faculty member has introduced new constructs of both
> YFP and CFP for localization in plant cells, mostly tobacco and
> arabidopsis. Our current system, a FluoView 500 is not set up for
> this work, so I've been asked to investigate new avenues.
>
> In any case, I think we've all seen a spectral system separate GFP,
> Sytox Green and FITC in the same cell, but for CFP, YFP, GFP, RFP
> colocalization in plant cells would a spectral system offer more
> utility than a system with better filter sets and laser lines?
>
> Obviously there is a cost question in relation to utility here. Also,
> here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon
> systems. If anyone would like comment on those in particular, that'd
> be great.
>
> I also have a more specific question, with several groups targeting
> chloroplasts, I should probably just lean towards a spectral system?
> Finally, do YFP and GFP overlap too greatly in plants (pH
> difference?!?) to be separated by either system?
>
> If any has some negative feedback, or suggestions, please feel free to
> email me privately. I find folks are awfully nice on the list, but
> when we're talking several hundred thousand dollars, I need brutal
> honesty.
>
> Thank you.
>
> Christian
>
--
Kurt Thorn, PhD
Director, Nikon Imaging Center
University of California San Francisco
UCSF MC 2140
Genentech Hall Room S252
600 16th St.
San Francisco, CA 94158-2517
http://nic.ucsf.eduphone 415.514.9709
fax 415.514.4300