http://confocal-microscopy-list.275.s1.nabble.com/FW-Cleaning-lens-tp2567899p2583166.html
maximum. Sadly the Hammamatsu Orca 100 [C4742-95]* is limited to a 10
seconds maximum exposure [and we really needed that 120 seconds]. All we
nowhere near enough for our Hammatsu Orca 100 camera to image anything. I
wasn't designed for very low light levels anyway]. Plus our inverted
> On your widefield microscope with camera, did you try binning of the
> pixels to image your luciferase?
>
> For BRET imaging tests using luciferase, where very high resolution was
> not important, we used a CoolSnapHQ binning 4 (then giving a 26 um pixel
> size) on the side port with 60 to 120 sec exposure time to see where was
> the signal. Best resolution have been done with a Cascade 512B camera
> (EMCCD backilluminated) with 30 sec exposure and binning 1. See the Julie
> Perroy article:
> .
> Biophys J. 2008 February 1; 94(3): 10011009.
>
> Subcellular Imaging of Dynamic Protein Interactions by Bioluminescence
> Resonance Energy Transfer
> Vincent Coulon,* Martin Audet, Vincent Homburger,* Joël Bockaert,*
> Laurent Fagni,* Michel Bouvier, and Julie Perroy*
>
>
> Monique Vasseur
> Microscopie et imagerie
> Département de biochimie
> Université de Montréal
> C.P. 6128, succursale Centre-ville
> Montréal QC H3C 3J7 Canada
> tél. (514) 343-6111 poste 5148
> -----Message d'origine-----
> De : Confocal Microscopy List [mailto:
[hidden email]] De
> la part de Robert J. Palmer Jr.
> Envoyé : 3 avril 2009 09:59
> À :
[hidden email]
> Objet : Re: Luciferase (Gaussia) activity detection
>
> You can do this with a true intensified
> photon-counting camera, e.g., those from
> Hamamatsu. Can't locate on their web site the
> model we used about 15 yrs ago, but it was based
> on a multi-channel plate that provided about 10E6
> amplification. We mounted it on a side port on
> our confocal and put the scope in a light box.
> The thing was sensitive enough to detect
> bacterial luciferase in single cells. As noted,
> lots of aluminum foil is recommend. I assume you
> are NOT trying to do three-D work, but I've
> always been intrigued by that possibility.
>
> At 9:39 AM -0400 4/3/09, Keith J Morris wrote:
>>No it probably isn't possible to use a confocal for luciferase. Just
>> tried
>>it with our standard wide field inverted microscope, all lights off, lots
>>of tin foil, Orca ER peltier cooled camera, maximum gain and not a hint
>> of
>>anything other than background noise. Our Orca ER has a 10 second max
>>exposure time, and the literature talks of 120 seconds plus exposure
>> times
>>[and the max camera gain didn't help us]. We concluded we would need a
>>£25k EMCCD camera or better to use our microscope with our samples, so
>> our
>>user is simply switching to an immunofluorescence method instead and
>>giving Luciferase a miss [just not worth spending the time developing the
>>method without the right microscope kit].
>>
>>A point scanning confocal will probably be less capable than our Orca ER
>>camera, and certainly no better. You loose too much light in the optics,
>>as there's so little to begin with, and so you need a very expensive
>>low-noise sensitive camera. Normally you use one of those large
>>light-tight cabinet type microscope imagers, bit like a plate scanner.
>>
>>Olympus do make the 'boxed' LV200 microscope for this type of work:
>>
>>
http://www.microscopy.olympus.eu/microscopes/Life_Science_Microscopes_Bioluminescence_Imaging_System_LV200_Applications.htm>>
>>and they talk of 15 minute exposure integration times for image capture.
>>If you have an expensive cooled B&W fluorescence microscope camera that
>>integrate for many minutes it might be worth a trying that though -
>>provided you can get the microscope stage into 'absolute darkness' [you
>>even have to get down to covering LEDs with insulating tape].
>>
>>Our user was using a luceriferase kit that reduced light levels a tad to
>>hopefully keep it going brighter for far longer than the few seconds
>>Luciferase normally lasts.
>>
>>Regards
>>
>>Keith
>>
>>-----------------------------------------------------------------
>>
>>Dr Keith J Morris
>>Molecular Cytogenetics and Microscopy Core
>>The Wellcome Trust Centre for Human Genetics
>>Roosevelt Drive
>>Oxford
>>OX3 7BN
>>United Kingdom
>>
>>
>>> Hi to all,
>>>
>>> Is it possible to detect luciferase activity on cell with a standard
>>> confocal micriscope (let's say LSM 510 and LSM 710 equipped with
>>> regular
>>> and spectral detector)?
>>>
>>> Emission of coelenterazine is at 470 nm.
>>>
>>> Thanks for you help!
>>>
>>> Louis
>>> Louis Villeneuve
>>> Research Associate- Confocal Microscopy
>>> Heart Montreal Institute- Research Center
>>> 5000 East Belanger
>>> Montreal (Qc), Canada
>>> H1T 1C8
>>>
>>> 514-376-3330 ext 3511
>>> 514-376-1355 (Fax)
>>>
>>>
[hidden email]
>
>
> --
> Robert J. Palmer Jr., Ph.D.
> Natl Inst Dental Craniofacial Res - Natl Insts Health
> Oral Infection and Immunity Branch
> Bldg 30, Room 310
> 30 Convent Drive
> Bethesda MD 20892
> ph 301-594-0025
> fax 301-402-0396
>