Re: Correction of Z-axis distortion- request for opinion

Posted by Stanislav Vitha on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Correction-of-Z-axis-distortion-request-for-opinion-tp2580156p2584556.html

I would also consider switching to a mounting medium with correct refractive
index (=same RI as immersion oil)  2,2-thiodiethanol comes to mind.  See
Staudt, 2007.  A 500 ml bottle can be had for ~$30.

1. Staudt, T., M.C. Lang, R. Medda, J. Engelhardt, and S.W. Hell, 2,2'-
thiodiethanol: a new water soluble mounting medium for high resolution optical
microscopy. Microsc Res Tech, 2007. 70(1): p. 1-9.

Stan Vitha
Microscopy and Imaging Center
Texas A&M University
 
On Fri, 3 Apr 2009 09:30:20 -0700, Armstrong, Brian <[hidden email]>
wrote:

>Michael, to be clear, your Z resolution will be approximately 3 times
>worse than your X,Y. Therefore, your Z will appear stretched. To address
>this you will need to perform deconvolution on your Z-Stacks. I would
>not worry too much the about RI of your oil and glycerin based mountant,
>as this is a very common procedure to use an oil imm objective on a
>glycerin based mounted sample and a glycerin imm objective will run
>around $10K US. I would however insist on a 1.5 coverslip that is 170um
>because your lens is designed for the angle created by that thickness
>and it is easy and cheap to use the right coverglass. After
>deconvolution I would recommend you use image software designed for
>accurate 3D visualization such as Imaris or Amira or Volocity.
>Cheers,
>
>Brian D Armstrong PhD
>Light Microscopy Core Manager
>Beckman Research Institute
>City of Hope
>Dept of Neuroscience
>1450 E Duarte Rd
>Duarte, CA 91010
>626-256-4673 x62872
>http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imag
>ing/Pages/default.aspx
>
>-----Original Message-----
>From: Confocal Microscopy List
[mailto:[hidden email]]

>On Behalf Of Michal Gdula
>Sent: Friday, April 03, 2009 4:21 AM
>To: [hidden email]
>Subject: Correction of Z-axis distortion- request for opinion
>
>Dear Confocalists,
>
>I need opinion of somebody experienced in correcting spherical
>aberrations to
>asses my approach.
>I am using Carl Zeiss LSM 510M microscope (with 63x apochromat
>objective,
>NA=1.4) and I noticed significant stretching in Z-axis.
>I found out from the literature that most probably it happens because of
>
>refraction index mismatch between immersion oil (Zeiss Immersol 518F
>ne=1.518) and mounting medium (Vectashield, Vector ne=1.457) or between
>refraction index of the specimen (20um thin skin cryosection processed
>according to FISH procedure). I have also discovered lately that we are
>supplied with the cover slips no. 1 with the thickness 130-160 um,
>whereas
>the optimal is 170um, however it is written in the Zeiss manual that
>immerse
>oil objective should be not sensitive to the differences in cover slip
>thikness.
>One of my aims is to measure  distances between FISH signals in 3D and I
>
>have to be  as accurate as possible.
>So far I  have done chromatic shift correction using measurements of
>differences between centroids of 0.5 um Tetraspeck beads in different
>channels.
>I am planning to measure z-axis distortion scanning 4um TetraSpeck beads
>
>and then calculate the differences between dimensions in the x,y and
>z-axis. I
>will prepare 2 slides with beads on the microscope slide and the second
>with
>the bead on the cover slip to check the difference of the aberration in
>different depths. This data will serve me to correct z-coordinates of
>FISH
>signals - I will calculate average ratio x-axis/z-axis and multiply the
>z-
>coordinates.
>I know that some scientists use some more sophisticated ways for
>correction
>of z-axis distortion. I would be grateful for any opinions and remarks.
>
>Best wishes,
>
>Michal Gdula
>Research PhD student
>[hidden email]
>Bradford University
>
>
>---------------------------------------------------------------------
>
>SECURITY/CONFIDENTIALITY WARNING:  
>This message and any attachments are intended solely for the individual or
entity to which they are addressed. This communication may contain
information that is privileged, confidential, or exempt from disclosure under
applicable law (e.g., personal health information, research data, financial
information). Because this e-mail has been sent without encryption, individuals
other than the intended recipient may be able to view the information, forward
it to others or tamper with the information without the knowledge or consent
of the sender. If you are not the intended recipient, or the employee or
person responsible for delivering the message to the intended recipient, any
dissemination, distribution or copying of the communication is strictly
prohibited. If you received the communication in error, please notify the
sender immediately by replying to this message and deleting the message and
any accompanying files from your system. If, due to the security risks, you do
not wish to receive further communications via e-mail, please reply to this
message and inform the sender that you do not wish to receive further e-mail
from the sender.  
>---------------------------------------------------------------------