Hi Young,
PFA fixation should not totally kill your GFP, it may weaken it
a bit but not totally wipe it out. If you find you have lost all your fluorescence
after fixation you can always go in with an anti-GFP antibody tagged with
Alexa488 or similar and get the signal back.
I have fixed numerous tissues and cells in 4% PFA and seen very
little degradation of the signal. For more aggressive staining (like BrdU incorporation
that requires acid fixation) the signal is gone, this is when the anti-GFP
antibody is useful.
Cheers
Cam