Hi Young,
Quantifying fluorescence intensity is always difficult, if not
impossible. You will be able to measure relative shifts in intensity between
samples providing they are all treated the same, stained the same and captured
using the same settings (eg. Exposure time or gain setting etc.).
PFA should be fine for all that. My recommendation is to
purchase PFA that is supplied in ampoules. We use a 16% PFA that is stored in
an ampoule under nitrogen gas. This is because PFA will oxidise to methanol
fairly quickly. This could be why you are seeing GFP loss with your fixation,
your fixative maybe all methanol by now.
Cheers
Cam
Cameron J. Nowell
Microscopy Manager
Centre for Advanced Microscopy
Ludwig Institute for Cancer Research
PO Box 2008
Royal Melbourne Hospital
Victoria, 3050
AUSTRALIA
Office: +61 3 9341 3155
Mobile: +61422882700
Fax: +61 3 9341 3104
From: Confocal Microscopy List
[mailto:[hidden email]] On Behalf Of Young Jik Kwon
Sent: Thursday, 23 April 2009 1:20 PM
To: [hidden email]
Subject: Re: GFP tissue preparation for confocal microscopy
Hi Cam,
Thanks for the information. We are trying to quantify gene expression by
fluorescence intensity. Do you think PFA would work for such work?
Best,
Young
On Wed, Apr 22, 2009 at 8:01 PM, Cameron Nowell <[hidden email]>
wrote:
Hi Young,
PFA fixation should not totally
kill your GFP, it may weaken it a bit but not totally wipe it out. If you find
you have lost all your fluorescence after fixation you can always go in with an
anti-GFP antibody tagged with Alexa488 or similar and get the signal back.
I have fixed numerous tissues
and cells in 4% PFA and seen very little degradation of the signal. For more
aggressive staining (like BrdU incorporation that requires acid fixation) the
signal is gone, this is when the anti-GFP antibody is useful.
Cheers
Cam
Cameron J. Nowell
Microscopy Manager
Centre for
Advanced Microscopy
Ludwig Institute
for Cancer Research
PO Box 2008
Royal Melbourne
Hospital
Victoria, 3050
AUSTRALIA
Office: +61 3 9341 3155
Mobile: +61422882700
Fax: +61 3 9341 3104
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Young Jik Kwon
Sent: Thursday, 23 April 2009 12:34 PM
To: [hidden email]
Subject: GFP tissue preparation for confocal microscopy
All,
We are trying to take confocal micrographs of at tumor tissues that express
GFP. What would be the sample preparation procedures? We tried cryosectioned
samples but the cell morphology seems weird. We already know paraformaldehyde
fixation kills GFP fluorescence. Any expert's advice?
Best,
Young
No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 8.5.287 / Virus Database: 270.12.2/2072 - Release Date: 04/22/09
17:25:00
This communication is intended only for the named recipient
and may contain information that is confidential, legally privileged or subject
to copyright; the Ludwig Institute for Cancer Research Ltd does not waiver any
rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not
necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.
No virus
found in this incoming message.
Checked by AVG - www.avg.com
Version: 8.5.287 / Virus Database: 270.12.2/2072 - Release Date: 04/22/09
17:25:00
Free forum by Nabble | Edit this page |