I totally agree with Steve: only bad PFA
kills GFP. I have done a fair bit of imaging GFP in filopodia which retracts
very fast if not fixed properly.
My understanding is that paraformaldehyde
is a polymer of formaldehyde and is solid. Formaldehyde is a gas that readily
dissolves in water.
In a formaldehyde solution, two things
happen:
- Formaldehyde equilibrates between the
air above the solution and the solution itself so every time you open your
bottle, more formaldehyde leaves the solution. Thus 15mL tubes are a good way
to store working stocks because the volume/surface ratio limits this process.
- The second thing that happens is that
paraformaldehyde forms as a precipitate which is definitely a sign that your
solution needs to be trashed (not down the sink of course!). Thus I think it is
not a good idea to buy ‘4%PFA’ in big bottles as it is usually sold
except maybe if you use it very fast or if you have applications where the
quality of fixation is not critical.
I prepare an 8% stock the following way:
Under a fume hood, warm up 150mL of PBS to
about 60ºC
Add 16g of PFA powder
Stir and keep warm to dissolve
Adjust the pH to 7.4 with concentrate NaOH
(you should then obtain a clear solution)
Adjust the volume to 200mL with PBS
Aliquote in 10 mL in 15mL falcon tubes
Store the aliquots for up to several years
(currently using a 2 years old stock) at -20º
The working stock can be stored at 4ºC for
a couple of months at the most
I normally pre warm the amount of solution
I need to 37ºC and keep the rest in the fridge
I add the warm 2x solution directly onto my
cells covered in the same volume of their own overnight medium (or changed at
least 30 min before sot that the cells are very happyJ)
5 min at 37ºC for a single layer culture
is plenty
Mount and image within 24h as the autofluorescence
will quickly show up against your signal. You can image later but your signal
to noise ratio will worsen with time.
I find this method excellent for
preserving filopodia. It’s also trouble free since there is no rinsing
the cells or making fresh solution every day.
Good luck!
Med vänlig hälsning / Best regards
Sylvie
@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader
Dept of Biosciences and Nutrition
Karolinska Institutet
Novum
14157
+46 (0)8 608 9240
From:
Sent: 23 April 2009 17:20
To:
[hidden email]
Subject: Re: GFP tissue
preparation for confocal microscopy
PFA does not kill GFP fluorescence.
BAD PFA kills GFP fluorescence.
We routinely use 1% PFA with GFP, YFP, and CFP, and have never had a problem so
long as the PFA is fresh and pH’ed to between 7.0 and 7.4.
-Steve
On 4/22/09 10:34 PM, "Young Jik Kwon" <[hidden email]> wrote:
All,
We are trying to take confocal micrographs of at tumor tissues that express
GFP. What would be the sample preparation procedures? We tried cryosectioned
samples but the cell morphology seems weird. We already know paraformaldehyde
fixation kills GFP fluorescence. Any expert's advice?
Best,
Young
****************************************************************************
Stephen C. Bunnell, Ph.D.
Assistant Professor
Department of Pathology
Jaharis Bldg., Room 512
Phone: (617) 636-2174
Fax: (617) 636-2990
Email: [hidden email]
SHIPPING ADDRESS (for packages):
Attn: Bunnell/Pathology/Jaharis 524
| Free forum by Nabble | Edit this page |