Re: GFP tissue preparation for confocal microscopy

Posted by Knecht, David on
URL: http://confocal-microscopy-list.275.s1.nabble.com/GFP-tissue-preparation-for-confocal-microscopy-tp2680696p2693589.html

We usually use 1-2% formaldehyde, often with 0.1% glutaraldehyde mixed in (also from ampule aliquots) in whatever media the cells are in.  This fixes slightly faster in some cells and the low glut does not generate significant background, so no quenching is necessary.  It is hard to quantify any of this, but this has become our "standard" fix as it has worked well for most cell types and probes.  Dave

On Apr 24, 2009, at 11:27 AM, Stephen Bunnell wrote:

I totally agree with Sylvie...  :)

Her chemistry is right on, and her PFA prep is basically identical to the one we use. My only addition is- Keep the powdered PFA stock at –20°C, and don’t open the stock while the bottle is still cold. This will lead to condensation into the stock, which promotes its degradation.

For preservation of fine/delicate/dynamic structures, we also use a final concentration of 4% PFA at 37°C to capture the ‘state’ instantly. We have added PFA to live cells in dynamic imaging assays, and it stops dynamic movements very quickly. If the PFA is good, you should not see _any_ fading of a GFP or YFP signal. If the PFA is good, leaving the cells in the fixative for ~24 hours does not appreciably reduce the YFP signal. Still, we usually wash it out after 10 minutes.

Some antibody epitopes (empirically determined) do not fare well under these conditions- In these cases we find that a longer fix with lower doses of PFA is adequate (e.g. 0.4% PFA for 30 minutes at 37°C). I suspect this approach would fail if the troublesome epitope were it to be found in a structure such as a filopodium.

    -Best regards,

    -Steve



On 4/24/09 4:25 AM, "Sylvie Le Guyader" <[hidden email]> wrote:

I totally agree with Steve: only bad PFA kills GFP. I have done a fair bit of imaging GFP in filopodia which retracts very fast if not fixed properly.
 
My understanding is that paraformaldehyde is a polymer of formaldehyde and is solid. Formaldehyde is a gas that readily dissolves in water.
In a formaldehyde solution, two things happen:
- Formaldehyde equilibrates between the air above the solution and the solution itself so every time you open your bottle, more formaldehyde leaves the solution. Thus 15mL tubes are a good way to store working stocks because the volume/surface ratio limits this process.
- The second thing that happens is that paraformaldehyde forms as a precipitate which is definitely a sign that your solution needs to be trashed (not down the sink of course!). Thus I think it is not a good idea to buy ‘4%PFA’ in big bottles as it is usually sold except maybe if you use it very fast or if you have applications where the quality of fixation is not critical.
 
I prepare an 8% stock the following way:
Under a fume hood, warm up 150mL of PBS to about 60ºC
Add 16g of PFA powder
Stir and keep warm to dissolve
Adjust the pH to 7.4 with concentrate NaOH (you should then obtain a clear solution)
Adjust the volume to 200mL with PBS
Aliquote in 10 mL in 15mL falcon tubes
Store the aliquots for up to several years (currently using a 2 years old stock) at -20º
The working stock can be stored at 4ºC for a couple of months at the most
I normally pre warm the amount of solution I need to 37ºC and keep the rest in the fridge
I add the warm 2x solution directly onto my cells covered in the same volume of their own overnight medium (or changed at least 30 min before sot that the cells are very happy
J)
5 min at 37ºC for a single layer culture is plenty
Mount and image within 24h as the autofluorescence will quickly show up against your signal. You can image later but your signal to noise ratio will worsen with time.
I find this method excellent for preserving filopodia. It’s also trouble free since there is no rinsing the cells or making fresh solution every day.
 
Good luck!
 

Med vänlig hälsning / Best regards

Sylvie

@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader
Dept of Biosciences and Nutrition
Karolinska Institutet
Novum
14157 Huddinge
Sweden
+46 (0)8 608 9240

From: Confocal Microscopy List [[hidden email] On Behalf Of Stephen Bunnell
Sent: 23 April 2009 17:20
To: [hidden email]
Subject: Re: GFP tissue preparation for confocal microscopy

PFA does not kill GFP fluorescence.

BAD PFA kills GFP fluorescence.

We routinely use 1% PFA with GFP, YFP, and CFP, and have never had a problem so long as the PFA is fresh and pH’ed to between 7.0 and 7.4.

    -Steve



On 4/22/09 10:34 PM, "Young Jik Kwon" <[hidden email]> wrote:
All,

We are trying to take confocal micrographs of at tumor tissues that express GFP. What would be the sample preparation procedures? We tried cryosectioned samples but the cell morphology seems weird. We already know paraformaldehyde fixation kills GFP fluorescence. Any expert's advice?

Best,

Young





****************************************************************************
Stephen C. Bunnell, Ph.D.
Assistant Professor
Tufts University Medical School
Department of Pathology
Jaharis Bldg., Room 512
150 Harrison Ave.
Boston, MA 02111

Phone: (617) 636-2174
Fax:   (617) 636-2990
Email: [hidden email]

SHIPPING ADDRESS (for packages):
Tufts University Receiving
37 Tyler St.
Attn: Bunnell/Pathology/Jaharis 524
Boston, MA 02111




****************************************************************************
Stephen C. Bunnell, Ph.D.
Assistant Professor
Tufts University Medical School
Department of Pathology
Jaharis Bldg., Room 512
150 Harrison Ave.
Boston, MA 02111

Phone: (617) 636-2174
Fax:   (617) 636-2990
Email: [hidden email]

SHIPPING ADDRESS (for packages):
Tufts University Receiving
37 Tyler St.
Attn: Bunnell/Pathology/Jaharis 524
Boston, MA 02111



Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)