Hi
Griffin,
Granted
the DAPI excitation is at the tail of the excitation peak*** when using the
405nm laser line, but there’s a heck of a lot of labeled DNA in the nucleus to
excite. The 405nm laser is essentially designed as a cheap replacement for an
incredibly expensive water cooled uV laser [£50k+ including cooling system] and
this 405nm line is still rarely used for anything other DAPI imaging. Hoechst
is
actually even worse than DAPI with the 405nm laser line, down from 11% to nearer
5% on the excitation peak tail.
In
every case I have encountered where the 405nm line laser is consistently very
poor at imaging it requires a long visit by the Zeiss or Leica engineer [for our
German confocal systems]*. Fortunately our confocal systems are on maintenance
contracts as generally an awful lot of DAPI associated hardware seems to been
replaced during ‘repair’ [e.g. lasers, AOTF & other optics]. This assumes
you have done the most a user can do, like align pinholes [for Zeiss 510s], etc…
I always have a perfectly DAPI/FITC/TRITC labeled Invitrogen BPAE cell and
kidney slice** to hand to either demonstrate to users that ‘the system is
working perfectly and it’s their poor samples’ or, more rarely, that ‘we have a
problem Huston’. The BPAE cells preps seem to store better than the kidney
slice, where in the latter the fluorescence diffuses to ‘incorrect’ locations
over time [it stays bright though]. These Invitrogen slides cost approx £80 and
£130 respectively and last up to a year stored in the
fridge/freezer.
As
other threads have pointed out, DAPI labeling isn’t always perfectly
straight-forward. I’d advise staining separately for DAPI and avoid the Likes of
‘Vectashield+DAPI’ to keep the background to the absolute minimum for top notch
DAPI images- although use of the latter shouldn’t produce really naff ‘off’ DAPI
images
Keith
*Assuming
it was decent at installation – and with our Zeiss 510 and Leica SP2 confocals
DAPI [and even Hoechst] imaging is excellent. We image at around 10% laser power
[30mW 405nm laser].
**http://www.well.ox.ac.uk/cytogenetics/Zeiss.shtml
see our Invitrogen slide
images
***http://www.invitrogen.com/site/us/en/home/support/Research-Tools/Fluorescence-SpectraViewer.reg.uk.html
---------------------------------------------------------------------------
Dr
Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory
00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt
Drive,
Oxford OX3 7BN,
United Kingdom.
Telephone: +44
(0)1865 287568
Email: [hidden email]
Web-pages: http://www.well.ox.ac.uk/cytogenetics/
From: Confocal Microscopy
List [mailto:[hidden email]] On Behalf Of Perry,
Griffin M
Sent: 23 June 2009 21:31
To:
[hidden email]
Subject: Nuclear Dyes at
405nm?
Dear
Listserv,
I was wondering if
anyone had better results with DAPI @ 405nm than I have. It
seems…off. I know it is only getting excited @ 25%, but I know another
group that gets good images at the same wavelength. The staining has been
uneven with DAPI, but with Topro-3 in the 638nm range, the staining has been
beautiful with clearly defined nucleoli. Does anybody use Hoechst
34580? Are there any other alternatives to DAPI @ the 405nm
range?
Thanks in
advance,
-Griffin
Internal Virus Database is out-of-date.
Checked by
AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date:
15/05/2009 6:16 AM
Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM
| Free forum by Nabble | Edit this page |