It's a good question and I think Shalin and Claire have given good
answers already. Perhaps there is a way to determine on the fly
during an experiment if you are indeed saturating the fluorescence
signal in any parts of the cell, but without that knowledge you could
be doing damage to the cells. Generally, it has been my experience,
and also it was emphasized many times at Jim Pawley's live-cell
microscopy course that I attended a few years ago, that the products
of photobleaching are toxic to the cell, and so using as little light
as possible is preferred so long as you can compromise with a longer
exposure time or increased detector gain and still get an acceptable
SNR for the phenomenon you're trying to observe. That being said,
there may be situations where you are trying to observe a fast
dynamic process in the cell by time-lapse imaging which necessitates
short exposure times and using a larger laser power to get a good SNR
again. As I understand it, some of the super-resolution imaging
techniques even depend on the non-linear effects of fluorescence
saturation to achieve imaging resolution below the diffraction limit,
so there are a few times when a larger laser power might be called for.
So, I would say there is no one definite recipe for all imaging
situations, but keeping the cells alive on the microscope stage is a
priority. Jennifer Waters has a very nice recent review on "Accuracy
and Precision in Light Microscopy" in the Journal of Cell Biology
where she rightly points out that it's up to the experimenter to
test and figure out the optimum settings of acquisition and perform
the proper control experiments to back up the interpretation results
of an imaging experiment.
John Oreopoulos
On 13-Jul-09, at 4:01 AM, Shalin Mehta wrote:
> Hi Sebastian,
>
> You are nearly correct in saying that the product of excitation power
> and exposure time (i.e. total energy put in the sample) is invariant.
> This linear relationship breaks down when you increase power too much,
> so that fluorophore gets 'saturated' from too many photons trying to
> excite it in short time.
>
> The energy (E) of the photon and wavelength (L) are related by
> Planck's constant (h) and speed of light (c): E=hc/L. The basic model
> of fluorescence is that the fluorophore absorbs a photon (with varying
> probabilities) across a range of wavelengths and after some time
> (called lifetime ) typically of pico to nano second values emits a
> lower energy photon.
>
> Ideally, you get one emission photon per excitation photon. When you
> put in certain power, you are imparting certain energy per second,
> i.e., putting in a given number of photons per second. As long as the
> time-rate of photons hitting the fluorophore is low you should get an
> emission photon per excitation photon. But at very high power ( high
> pohton-rates) this relationship becomes non-linear, and you get less
> number of emission photons per excitation photon.
>
> You may find introductory chapters of "Principles of fluorescence
> spectroscopy" by Joseph Lakowicz illuminating in this regard.
>
> best
> shalin
>
> On Mon, Jul 13, 2009 at 2:18 AM, Sebastien
> Stephens<
[hidden email]> wrote:
>> Hi,
>>
>> 100% laser for 1 ms OR 1% laser for 100ms?
>>
>> This is a subject that has been bothering me and I have never got
>> quite the
>> answer for it.
>>
>> After speaking to someone who uses 100% laser power (sorry I don’t
>> know
>> what the actual laser output is) I noted their goal was to expose
>> for the least
>> amount of time possible (which was only a few millesecs). This
>> enticed me to
>> check myself how my protein structures (podosomes) could handle
>> 100% laser
>> power as opposed to 30% which I usually use. As expected the exposure
>> times went down significantly (from 320 ms to 30ms) and in fact I
>> was able to
>> still get decent information from 10ms exposures taken every sec
>> for 3 mins.
>> Surprisingly, I saw no detectable bleach and it did not seem any
>> different to
>> long exposure times with less laser power.
>>
>> Im only new to microscopy and surely others must have experimented
>> with:
>> high laser+ low exp compared to low laser+high exp.
>>
>> I really want to know, which one should I do (high or low power)?
>>
>> It seems that there could be mathematically and practically
>> equivalents of
>> high laser+low exp to low laser+high exp (ie as above 100ms 1%
>> power pairs
>> with 1ms 100% laser power). So, is there a relationship between
>> power and
>> exposure time exists that can sort of be mathematically related?
>>
>> Cheers
>>
>> Seb
>>