Re: resolution target

Posted by Rosemary.White on
URL: http://confocal-microscopy-list.275.s1.nabble.com/resolution-target-tp3298871p3300960.html

Re: resolution target Another way to test optics is to coat a clean, but not ultraclean, glass slide with a reasonable amount of gold (i.e. thicker than you’d normally use for SEM) in a standard sputter coater, then wipe the coated surface very gently.  This will produce tiny pinholes, just select the size you want to use.  To preserve the surface, glue a long coverslip on top with your preferred mountant, that way you have the correct amount of glass between objective and object.  It’s a good way of testing chromatic aberration, for example.

cheers,
Rosemary


Rosemary White
CSIRO Plant Industry
GPO Box 1600
Canberra, ACT 2601
Australia

ph 61 2 6246 5475
fx 61 2 6246 5334


On 22/07/09 1:49 AM, "Julio Vazquez" <jvazquez@...> wrote:

Hi Thomas,

Mr Klaus Kemp in the UK used to make diatom slides, including one with 8 species of diatoms covering a broad range of resolutions. The slides were about 20 Euros or so. I am not sure whether he is still in business though. Below is his web site, and the list of diatoms on his 8-form test plate.

Julio.

http://www.diatoms.co.uk/

  • Amphipleura pellucida 270 nm
  • Frustulia rhomboides 300 nm
  • Pleurosigma angulatum 525 nm
  • Surirella gemma 500 nm
  • Nitzschia sigma 440 nm
  • Stauroneis phoenicentron 720 nm
  • Navicula lyra 1250 nm
  • Gyrosigma balticum 670 nm

There also used to be an article by Dave Walker: Counting the dots: Giving Microscopes a workout using diatom test slides at  
www.microscopy-uk.org.uk/mag/artoct99/dwdiatom.html    

the article has been inaccessible for a while though


 
--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024


http://www.fhcrc.org

==


On Jul 21, 2009, at 5:11 AM, Thomas Aabo wrote:

Hi
Does anyone know a good (cheap) place to get a positive high resolution target for a bright field microscope (50X, N.A. 0.55). Inside the EU would be best.
Thanks
Thomas

-------------------------------------------------------------