Stephen: I agree, technically 405 is 'violet' not UV. At best you
terms of raw power transmission through the objective. Your mileage
> G'day Renato,
>
> I'm not sure I would describe 405nm as UV, I tend to call it visible violet.
> 405nm should be transmitted very well by most objectives. Transmission
> curves for Olympus objectives can be found at
>
http://microscope.olympus.com/uis2/ . I could not find any objective listed
> there with poor transmission at 405nm.
>
> When doing Ca2+ imaging with Fura-2 a combination of 340nm and 380nm is
> often used for excitation and certainly in order to transmitt 405nm
> efficiently a UV objective should be used. But if no UV transmitting
> objectives are available it ispossible to use 360nm and 380nm combination
> instead. You should not have problems with Visible Violet transmission at
> 405nm with any objective.
>
> Having said that some objectives are chormatically corrected at 405nm, which
> can improve imaging especially when using confocal.
> Cheers
> Stephen Cody
>
> 2009/9/22 Renato Mortara <
[hidden email]>
>>
>> Hello,
>>
>> I wonder if I could get some feedback on how two doubts;
>>
>> 1) How well does the solid-state 405 laser performs with DAPI-labeled
>> samples (Cells, tissues, etc.) with not-so-UV-transparent objectives ?
>>
>> I currently run an old BioRad 1024UV and with the Coherent 622 Inova laser
>> (351/363 lines) even with not-so-UV-transparent objectives such as a 63x
>> PlanApo and a 100X 1.4 PlanApo (both from Zeiss), by using DAPI at a
>> relatively high concentration (20-50 uM) we get very nice signal.
>>
>> 2) Has anyone tried the 150x 1.45 NA Olympus objective (presumably
>> designed for TIRF) with conventional/confocal fluorescence ? Is the small WD
>> a problem ?
>>
>>
>> Any input will be greatly appreciated,
>>
>> Best regards,
>> Renato
>>