Re: Eyepiece and virtual image
Hi all,
This topic causes lots of confusion at the UBC Live Cell
Course.
We try to emphasize three things:
1. That any lens will create a real image of any object on its
"front" side at some distance on its "back" side
as long as the object is more than one focal length away from the
optical center of the lens.
2. The magnification of this image will be directly proportional
to the ratio of the distance between the central plane of the lens and
the image divided by the distance between the object and the central
plane. (i.e, With any converging lens, you can get any
magnification as long as you are flexible about where your object and
your images planes are.)
3. Therefore, the three reasons for using different lenses for
different magnifications are
a. by
having different focal lengths, these objectives can produce real
images of specific sizes (magnifications) without having to change
the object (specimen) and the image (CCD) planes.
b. Low
mag (i.e., long focal length) objectives can be located with their
optical center farther from the object and be optimized to cover
larger areas of the object while high magnification (i.e., shorter
focal length) objectives can be mounted with their optical centers
closer to the object and optimized to be free from aberrations at
higher numerical aperture.
c. Most
important of all: the aberrations inherent in all simple,
refractive lenses can only be compensated for at one particular set of
object and image planes.
So yes, an infinity objective (plus its tube lens but either in
the presence or the absence of an occular) will form real images of
the structures found at any plane in the object that is more than one
focal length beyond its optical center. The location of this image
along the optical axis will determine what plane in the object will be
in focus there.
BUT such an image will only be free from aberrations as
long as it is focused on one plane in the object. For
infinity-conjugate objectives, the ideal object plane is that found at
the front focal plane of the objective and it will focus light from
this plane into an (almost) perfect "intermediate" image on
a plane 1 cm below the lip of the tube holding the ocular.
The ocular itself is located so that its optical center is one
focal length above this intermediate image plane meaning that this
image emerges from the ocular as though it were coming from an
object an infinite distance away. This is important because, on a
binocular viewing unit, the two ocular tubes are parallel to each
other and so you would like the optical axes of your eyes to also be
parallel, a condition that will only occur when your eyes think that
they are focused on distant object. (The idea that this set up was
designed to allow one superimpose the image from the microscope with
that of an image being drawn by hand about one meter away is a useful
approximation, but isn't quite true) .
Yes, you can find real images of any plane in the object
projected onto some plane above the "eye point" or
"exit pupil" of the ocular, (or if you remove the ocular, at
some plane down the tube) but any such image will only be free from
aberrations if the image plane is at infinity (Which is why one uses a
special projection lens rather than an ocular to project the image
onto, for instance, a CCD camera).
I hope that this helps,
Cheers,
Jim Pawley
Yes--what you describe is quite true--but
the point is that you do have to refocus the microscope to make the
projected image sharp. In fact, there are old methods for using that
to take photos with plate-film cameras, where the image was refocused
on the camera's ground-glass screen.
But I think the answer to your original
question is that the intermediate image is not quite at the front
focal point of the eyepiece, so the eyepiece isn't really projecting
an image to infinity. In fact, when you look in the eyepiece, the
apparent (virtual) image distance is roughly 25cm away. My
understanding is that this is (historically speaking) so that you
could look in your upright monocular microscope with your left eye and
at your drawing paper on the table next to the microscope with your
right eye.
That said, I think the physical distance between the actual front
focal plane (image at infinity) and the intermediate image (image at
25 cm) is probably pretty small. I'm not able to walk through the math
to get there....
Julian
Joachim Hehl wrote:
I would say it is in focus. But yes it
gets a bit sharper when adjusting the focus. And the wider I go back
with the paper the bigger the image is, like a projector. Maybe you
try it out with one of your scopes?
Joachim
On [DATE], "Guy Cox" <[ADDRESS]> wrote:
But is this image truly in focus? If you adjust the
microscope
focus, does it get sharper?
Guy
------------------------------------------------------------------------
*From:* Confocal Microscopy List
[mailto:[hidden email]] *On
Behalf Of *Joachim Hehl
*Sent:* Thursday, 8 October 2009 8:03 PM
*To:* [hidden email]
*Subject:* Re: Eyepiece and virtual image
Dear Guy,
The point is that I neither adjust the focus nor
lift the
eyepiece. I look through the eyepiece (without my
glasses, I am
short-sighted), focusing and then I only hold a
piece of paper in
front of the eyepiece and can see the picture on
it. And this also
happens to my normal-sighted colleagues.
Joachim
On [DATE], "Guy Cox" <[ADDRESS]>
wrote:
You can always get a real
image from an eyepiece by
refocussing so that the
first image is in front of the focal
plane of the eyepiece -
either adjust the microscope focus a
little or lift the eyepiece
slightly in its tube. If you see a
sharp image without
refocussing from your normal viewing
position, it probably means
that you - like me - are long-sighted!
Guy
Optical Imaging Techniques
in Cell Biology
by Guy Cox CRC Press /
Taylor & Francis
http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy
Cox, MA, DPhil(Oxon)
Electron Microscope Unit,
Madsen Building F09,
University of Sydney, NSW
2006
______________________________________________
Phone +61 2 9351 3176 Fax
+61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net
<http://www.guycox.net/>
------------------------------------------------------------------------
*From:* Confocal Microscopy
List
[mailto:[hidden email]] *On Behalf Of
*Joachim Hehl
*Sent:* Thursday, 8 October
2009 7:36 PM
*To:*
[hidden email]
*Subject:* Eyepiece and
virtual image
Dear all,
I have a -maybe stupid-
and not confocal but "basic optic"
question:
As you can read in all
textbooks concerning microscopy and
geometric optics the
objective produces a real, inverted and
magnified image since the
distance from the object to the
object front lens is bigger
than one but less than two focal
lengths of that given
objective.
This intermediate image is
then magnified by the eyepiece.
Since the intermediate
image lies exactly in the front focal
plane of the eyepiece the
result is a virtual, true sided and
magnified image which occur
in the infinite space. Our eye
with its optical components
is then producing a real image on
the retina.
So far, so good.
By definition, a virtual
image can not be captured on a
screen. BUT: When I hold a
piece of paper in front of the
eyepiece in a distance
bigger or smaller than the back focal
plane of the eyepiece (the
distance I use when I look through
it with my eyes) I am able
to capture a pretty sharp image of
my object on the paper. Why
is this? I should not since it is
a virtual image?
Thanks for your input!
Joachim
Joachim Hehl
LMC-Light Microscopy
Centre, ETH Zurich Hönggerberg
Schafmattstrasse 18, HPM
F16.1
CH-8093, Zurich, Switzerland
Web: www.lmc.ethz.ch
Phone: +41 44 633 6202
Natel: +41 44 658 1679
Fax: +41 44 632 1298
e-mail:
[hidden email]
--
Julian P.S. Smith III
Director, Winthrop Microscopy Facility
Dept. of Biology
Winthrop University
520 Cherry Rd.
Rock Hill, SC 29733
803-323-2111 x6427 (vox)
803-323-3448 (fax)
803-524-2347 (cell)
--
James and Christine Pawley, 21 N. Prospect Ave. Madison, WI,
53726 Phone: 608-238-3953