Shigeo,

I'm not sure if the unusual polarization properties of TIRF  
microscopy should be called a "problem". As I said at the end of my  
last posting on this topic, polarized TIRF microscopy/spectroscopy  
can be exploited to assess the orientation or order of fluorescent  
molecules near a surface, and in that case I would consider the  
effect a benefit, not a problem.

On the other hand, if you're talking about single-molecule imaging  
(in vivo or in vitro), the polarization in TIRF is something to be  
aware of and perhaps controlled depending on what information is  
trying to be sought.

Fluorescing molecules behave like tiny electric antenna and the  
absorption of light by a single fluorescing molecule is polarization  
dependent. The origin of this behavior is based on the existence of a  
definite transition dipole moment vector for the absorption and  
emission of light that lies along a specific direction within the  
fluorophore structure. Not surprisingly, these transition dipole  
moments usually lie roughly along the chain of conjugated double  
bonds in the chemical structure which possess outer orbital electrons  
that can easily oscillate along these directions. This region of the  
molecule is called the chromophore. The PROBABILITY of light  
absorption (and subsequent fluorescence emission) by a single  
molecule is maximized when the polarization of the incident light  
(ie: the electric field vector associated with the light) is parallel  
to the transition dipole moment vector of the molecule and follows a  
cosine squared dependence for all other angles between the two  
vectors. If you'd like to see a macroscopic example of this with real  
electric antenna, check out the youtube video provided by the physics  
department at MIT:

http://www.youtube.com/watch?v=nCAKQQjfOvk&NR=1

Isolated fluorescing molecules behave the same way as the macroscopic  
radio antenna, and so the physics is the same. Lakowicz's book  
Principles of Fluorescence Spectroscopy contains a full mathematical  
derivation of the effect at the single-molecule level and explains  
how it is used in fluorescence spectroscopy applications. What is  
truly amazing is that this phenomenon can even be observed directly  
with an epifluorescence microscope equipped with a polarizing optic  
on the illumination side. See this paper:

Schutz, G. J., H. Schindler, and T. Schmidt. 1997. Imaging single-
molecule dichroism. Opt. Lett. 22:651-653.

Again, what is unique about polarized TIRF microscopy compared to  
polarized epifluorescence microscopy is that it is possible to  
preferentially excite single molecules that have their transition  
dipole moment oriented more vertically outside of the xy plane of  
imaging because the "p" polarization of TIRF is directed mostly along  
the z-direction (perpendicular to the imaging plane). Therefore, it  
becomes possible through some clever choices of modulated polarized  
excitation in TIRF to assess the full 3D orientation of a fluorescent  
molecule (on a given time-scale determined by the exposure time of  
imaging) relative to the sample substrate (the xy imaging plane  
again). Truly, the best example of this is the work undertaken  
earlier this decade by the Goldman group who used polarized TIRF  
illumination (and observation of the polarized emission) of singly  
fluorescently labeled Myosin molecular motors to assess the protein's  
structural dynamics on actin filaments. I am always astounded by  
complexity and ingenuity of these experiments as well as the  
information about molecular orientation that can be gleaned from them:

Forkey, J. N., M. E. Quinlan, and Y. E. Goldman. 2000. Protein  
structural dynamics by single-molecule fluorescence polarization.  
Prog. Biophys. Mol. Biol. 74:1-35.

Forkey, J. N., M. E. Quinlan, and Y. E. Goldman. 2005. Measurement of  
single macromolecule orientation by total internal reflection  
fluorescence polarization microscopy. Biophys. J. 89:1261-1271.

Note that the equipment used in the above examples is home-built and  
customized.
So in the cases stated above, I would not call the polarization of  
TIRF a problem, but rather an advantage. If on the other hand you are  
trying to measure some other property other than orientation of  
single molecules, then it might be considered a problem. For example,  
some researchers try observe the diffusion of single fluorescent  
molecules ("single particle tracking") to assess structural changes  
in the environment of the probe. Other researchers may try to observe  
FRET between two single fluorophores attached to a protein to assess  
dynamic changes in protein folding conformation at the single protein  
level ("single molecule intramolecular FRET"). In carefully  
calibrated single molecule experiments, it is even possible to count  
the number of molecules in a diffraction limited spot to assess  
absolute concentrations of labeled proteins, DNA, etc. ("number and  
brightness analysis"). All of these types of single-molecule  
measurements are intensity-based just like the single-molecule  
polarization measurements stated above, however. On top of that,  
single molecules sometimes exhibit some complicated photophysics that  
lead to so-called blinking on different time scales. So if you are  
trying perform one of these other types of single molecule  
measurements, it would be best to remove all polarization bias of the  
excitation of the molecule, especially if the molecules under study  
rapidly rotate or re-orient in time. Obviously, one way to do this  
would be to "depolarize" the incident illumination light used in your  
experiment. Unfortunately it is surprisingly difficult to create a  
perfectly depolarized source of light in a microscope since every  
time the light transmits through or reflects from an internal optic  
(lenses, mirrors, etc.) the light becomes polarized a small amount in  
one direction. In addition, lasers usually emit strongly linearly  
polarized light as well. You can check the polarization direction of  
your laser beam by rotating a film polarizer (polaroid) in the path  
of the beam, the direction of polarization being the the angle of the  
polaroid that gives you maximum transmission. Again, it is very  
difficult to depolarize the laser light and so the solution to the  
polarization "problem" in these cases is to create CIRCULARLY  
polarized light using an optic called a quarter-wave plate.  
Circularly polarized light can be considered light composed of an  
equal amount "s" and "p" polarized light with a 90 degree phase shift  
between these two electric field vectors. To learn more about this,  
see these links:

http://en.wikipedia.org/wiki/Polarized
http://en.wikipedia.org/wiki/Quarter-wave_plate

As you know now, "p" polarization in TIRF contains light partially  
polarized along the x-direction, but mostly along the z-direction.  
"s" polarized light in TIRF is purely along the y-direction. So  
circularly polarized light used in TIRF microscopy gives you  
polarized illumination that is somewhat isotropic or unbiased, but  
not perfectly so. You will sometimes (but not always) see mentioned  
in the methods sections of single-molecule study publications the  
addition of a quarter-wave plate in the laser illumination beam path.  
For one example, see the final paragraph of the following paper:

Toprak, E., J. Enderlein, S. Syed, S. A. McKinney, R. G. Petschek, T.  
Ha, Y. E. Goldman, and P. R. Selvin. 2006. Defocused orientation and  
position imaging (dopi) of myosin v. Proc. Natl. Acad. Sci. U. S. A.  
103:6495-6499.

Now I think I am in a good position to properly answer your two  
questions:

1. Most commercial TIRF microscope systems utilize laser illumination  
(with the exception of the so-called "white-light" TIRF systems that  
use a Mercury lamp), and so the incident light will be linearly  
polarized along a certain direction depending on the rotation angle  
of the laser cavity relative to the microscope entry port. As far as  
I know, the current commercial systems make no attempt to control  
this, so yes, in a single-molecule experiment (and depending on the  
orientational dynamics of the molecules under study) you might be  
preferentially selecting/exciting a sub-population of molecules that  
eventually emit fluorescence.

2. I believe that current commercial TIRF microscope systems only  
utilize a single incident laser beam directed through a single point  
on the periphery of a high NA microscope objective and not a ring-
shaped illumination pattern that encompasses the entire periphery of  
the objective aperture. I am aware of a few examples of these "ring-
beam" TIRF systems in the literature, but they are all home-built  
setups that "solve" this polarization problem. I have also seen  
examples of "flying-spot" TIRF illumination where the single beam is  
forced to rotate rapidly along the periphery leading to the same  
unbiased polarization effect that a ring-beam provides. These types  
of illumination are difficult to align and would only be of interest  
to those who are concerned with very precise single-molecule  
measurements where polarization bias matters. It is for these reasons  
that I suspect the commercial vendors choose to work with systems  
that utilize only a single illumination beam, but even here you have  
no control over the illumination polarization unless you insert your  
own polarization optics into the optical train. This not easy to do  
on a commercial system and I have had experience doing this on a home-
built TIRF system which is much more forgiving when it comes to  
inserting additional optics.

I think the key point here is that polarization bias in fluorescence  
imaging (epifluorescence or TIRF or even confocal) is a concern to  
only a select group of researchers, mostly in the single-molecule/
biophysics community. In most cell biology applications, the  
researcher concentrates on imaging the location of a labeled  
structure in a cell, tracking it in time, and perhaps looking for  
colocalization in two different channels. In most cases, the  
structure of interest will be labeled at a concentration such that  
hundreds or thousands of molecules are attached to it with a RANDOM  
orientation and the polarization of illumination light won't matter  
(no linear dichroism will be observed). But even here I should stress  
the phrase "most cases" since there have been reports on the  
listserver about polarization effects, especially for membrane  
probes. Sometimes, a fluorescence image can look quite different  
depending on the polarization of illumination. Dan Axelrod has  
already shown how membrane blebbing or invagination can be imaged  
using polarized TIRF illumination:

Sund, S. E., J. A. Swanson, and D. Axelrod. 1999. Cell membrane  
orientation visualized by polarized total internal reflection  
fluorescence. Biophys. J. 77:2266-2283.

John Oreopoulos


On 24-Dec-09, at 2:12 AM, Shigeo Watanabe wrote:

>
> Dear All,
>
> I have previouly asked how polarized the P- and S-polarized light  
> are at TIRF illumination.
> Thanks to John, now I get to know that P-polarized light is  
> converted to "cartwheel" polarized light while S-polized light is  
> intact.
> As he suggested I read the review of Dan Axelrod about the effect
> (or problem) of this cartwheel polized light when observing the  
> sample.
> What I understand in his review is that cartwheel light excites  
> only molecules which is parallel to z-axis.
>
> Researchers who I talked with about TIRF problem also mentioned  
> that normal TIRF system which use single incident light from one  
> entering direction excite only a fraction of  molecules which is  
> paralle to poloized evanescent light direction and then they prefer  
> to use the ring-like illumination for TIRF to excite every single  
> molecules.
>
> Now I am confused about the actual TIRF problem.
> Quesitons I have are these.
> 1)Does evanescent light excite the only molecules which are paralle  
> to polarization of evanescent light because of cartwheel polarized  
> light even when incident light is non-polarized light???
> I am wondering what the actual polarization of evanescent light  
> produced by non-polarized incident light, which is mixture of P-  
> and S- polarized light.
>
> 2)Do commercial TIRF systems have this polarized problem? Do they  
> use single incident light or ring-like incident light?
>
>
> I appreciate if anyone help to answer these questions.
>
> Sincerely
> Shigeo Watanabe
> Hamamatsu Photonics KK
>
>
>
>
>
>
>
>
>