Re: TIRF and p- s-polarized incident linght

Posted by John Oreopoulos on
URL: http://confocal-microscopy-list.275.s1.nabble.com/paper-help-tp4115966p4232515.html

Sorry, I sent the wrong Youtube link for the MIT video. See this one  
instead:

http://www.youtube.com/watch?v=4xF1Fq2wB1I&feature=channel

John

On 30-Dec-09, at 9:46 AM, John Oreopoulos wrote:

> Shigeo,
>
> I'm not sure if the unusual polarization properties of TIRF  
> microscopy should be called a "problem". As I said at the end of my  
> last posting on this topic, polarized TIRF microscopy/spectroscopy  
> can be exploited to assess the orientation or order of fluorescent  
> molecules near a surface, and in that case I would consider the  
> effect a benefit, not a problem.
>
> On the other hand, if you're talking about single-molecule imaging  
> (in vivo or in vitro), the polarization in TIRF is something to be  
> aware of and perhaps controlled depending on what information is  
> trying to be sought.
>
> Fluorescing molecules behave like tiny electric antenna and the  
> absorption of light by a single fluorescing molecule is  
> polarization dependent. The origin of this behavior is based on the  
> existence of a definite transition dipole moment vector for the  
> absorption and emission of light that lies along a specific  
> direction within the fluorophore structure. Not surprisingly, these  
> transition dipole moments usually lie roughly along the chain of  
> conjugated double bonds in the chemical structure which possess  
> outer orbital electrons that can easily oscillate along these  
> directions. This region of the molecule is called the chromophore.  
> The PROBABILITY of light absorption (and subsequent fluorescence  
> emission) by a single molecule is maximized when the polarization  
> of the incident light (ie: the electric field vector associated  
> with the light) is parallel to the transition dipole moment vector  
> of the molecule and follows a cosine squared dependence for all  
> other angles between the two vectors. If you'd like to see a  
> macroscopic example of this with real electric antenna, check out  
> the youtube video provided by the physics department at MIT:
>
> http://www.youtube.com/watch?v=nCAKQQjfOvk&NR=1
>
> Isolated fluorescing molecules behave the same way as the  
> macroscopic radio antenna, and so the physics is the same.  
> Lakowicz's book Principles of Fluorescence Spectroscopy contains a  
> full mathematical derivation of the effect at the single-molecule  
> level and explains how it is used in fluorescence spectroscopy  
> applications. What is truly amazing is that this phenomenon can  
> even be observed directly with an epifluorescence microscope  
> equipped with a polarizing optic on the illumination side. See this  
> paper:
>
> Schutz, G. J., H. Schindler, and T. Schmidt. 1997. Imaging single-
> molecule dichroism. Opt. Lett. 22:651-653.
>
> Again, what is unique about polarized TIRF microscopy compared to  
> polarized epifluorescence microscopy is that it is possible to  
> preferentially excite single molecules that have their transition  
> dipole moment oriented more vertically outside of the xy plane of  
> imaging because the "p" polarization of TIRF is directed mostly  
> along the z-direction (perpendicular to the imaging plane).  
> Therefore, it becomes possible through some clever choices of  
> modulated polarized excitation in TIRF to assess the full 3D  
> orientation of a fluorescent molecule (on a given time-scale  
> determined by the exposure time of imaging) relative to the sample  
> substrate (the xy imaging plane again). Truly, the best example of  
> this is the work undertaken earlier this decade by the Goldman  
> group who used polarized TIRF illumination (and observation of the  
> polarized emission) of singly fluorescently labeled Myosin  
> molecular motors to assess the protein's structural dynamics on  
> actin filaments. I am always astounded by complexity and ingenuity  
> of these experiments as well as the information about molecular  
> orientation that can be gleaned from them:
>
> Forkey, J. N., M. E. Quinlan, and Y. E. Goldman. 2000. Protein  
> structural dynamics by single-molecule fluorescence polarization.  
> Prog. Biophys. Mol. Biol. 74:1-35.
>
> Forkey, J. N., M. E. Quinlan, and Y. E. Goldman. 2005. Measurement  
> of single macromolecule orientation by total internal reflection  
> fluorescence polarization microscopy. Biophys. J. 89:1261-1271.
>
> Note that the equipment used in the above examples is home-built  
> and customized.
> So in the cases stated above, I would not call the polarization of  
> TIRF a problem, but rather an advantage. If on the other hand you  
> are trying to measure some other property other than orientation of  
> single molecules, then it might be considered a problem. For  
> example, some researchers try observe the diffusion of single  
> fluorescent molecules ("single particle tracking") to assess  
> structural changes in the environment of the probe. Other  
> researchers may try to observe FRET between two single fluorophores  
> attached to a protein to assess dynamic changes in protein folding  
> conformation at the single protein level ("single molecule  
> intramolecular FRET"). In carefully calibrated single molecule  
> experiments, it is even possible to count the number of molecules  
> in a diffraction limited spot to assess absolute concentrations of  
> labeled proteins, DNA, etc. ("number and brightness analysis"). All  
> of these types of single-molecule measurements are intensity-based  
> just like the single-molecule polarization measurements stated  
> above, however. On top of that, single molecules sometimes exhibit  
> some complicated photophysics that lead to so-called blinking on  
> different time scales. So if you are trying perform one of these  
> other types of single molecule measurements, it would be best to  
> remove all polarization bias of the excitation of the molecule,  
> especially if the molecules under study rapidly rotate or re-orient  
> in time. Obviously, one way to do this would be to "depolarize" the  
> incident illumination light used in your experiment. Unfortunately  
> it is surprisingly difficult to create a perfectly depolarized  
> source of light in a microscope since every time the light  
> transmits through or reflects from an internal optic (lenses,  
> mirrors, etc.) the light becomes polarized a small amount in one  
> direction. In addition, lasers usually emit strongly linearly  
> polarized light as well. You can check the polarization direction  
> of your laser beam by rotating a film polarizer (polaroid) in the  
> path of the beam, the direction of polarization being the the angle  
> of the polaroid that gives you maximum transmission. Again, it is  
> very difficult to depolarize the laser light and so the solution to  
> the polarization "problem" in these cases is to create CIRCULARLY  
> polarized light using an optic called a quarter-wave plate.  
> Circularly polarized light can be considered light composed of an  
> equal amount "s" and "p" polarized light with a 90 degree phase  
> shift between these two electric field vectors. To learn more about  
> this, see these links:
>
> http://en.wikipedia.org/wiki/Polarized
> http://en.wikipedia.org/wiki/Quarter-wave_plate
>
> As you know now, "p" polarization in TIRF contains light partially  
> polarized along the x-direction, but mostly along the z-direction.  
> "s" polarized light in TIRF is purely along the y-direction. So  
> circularly polarized light used in TIRF microscopy gives you  
> polarized illumination that is somewhat isotropic or unbiased, but  
> not perfectly so. You will sometimes (but not always) see mentioned  
> in the methods sections of single-molecule study publications the  
> addition of a quarter-wave plate in the laser illumination beam  
> path. For one example, see the final paragraph of the following paper:
>
> Toprak, E., J. Enderlein, S. Syed, S. A. McKinney, R. G. Petschek,  
> T. Ha, Y. E. Goldman, and P. R. Selvin. 2006. Defocused orientation  
> and position imaging (dopi) of myosin v. Proc. Natl. Acad. Sci. U.  
> S. A. 103:6495-6499.
>
> Now I think I am in a good position to properly answer your two  
> questions:
>
> 1. Most commercial TIRF microscope systems utilize laser  
> illumination (with the exception of the so-called "white-light"  
> TIRF systems that use a Mercury lamp), and so the incident light  
> will be linearly polarized along a certain direction depending on  
> the rotation angle of the laser cavity relative to the microscope  
> entry port. As far as I know, the current commercial systems make  
> no attempt to control this, so yes, in a single-molecule experiment  
> (and depending on the orientational dynamics of the molecules under  
> study) you might be preferentially selecting/exciting a sub-
> population of molecules that eventually emit fluorescence.
>
> 2. I believe that current commercial TIRF microscope systems only  
> utilize a single incident laser beam directed through a single  
> point on the periphery of a high NA microscope objective and not a  
> ring-shaped illumination pattern that encompasses the entire  
> periphery of the objective aperture. I am aware of a few examples  
> of these "ring-beam" TIRF systems in the literature, but they are  
> all home-built setups that "solve" this polarization problem. I  
> have also seen examples of "flying-spot" TIRF illumination where  
> the single beam is forced to rotate rapidly along the periphery  
> leading to the same unbiased polarization effect that a ring-beam  
> provides. These types of illumination are difficult to align and  
> would only be of interest to those who are concerned with very  
> precise single-molecule measurements where polarization bias  
> matters. It is for these reasons that I suspect the commercial  
> vendors choose to work with systems that utilize only a single  
> illumination beam, but even here you have no control over the  
> illumination polarization unless you insert your own polarization  
> optics into the optical train. This not easy to do on a commercial  
> system and I have had experience doing this on a home-built TIRF  
> system which is much more forgiving when it comes to inserting  
> additional optics.
>
> I think the key point here is that polarization bias in  
> fluorescence imaging (epifluorescence or TIRF or even confocal) is  
> a concern to only a select group of researchers, mostly in the  
> single-molecule/biophysics community. In most cell biology  
> applications, the researcher concentrates on imaging the location  
> of a labeled structure in a cell, tracking it in time, and perhaps  
> looking for colocalization in two different channels. In most  
> cases, the structure of interest will be labeled at a concentration  
> such that hundreds or thousands of molecules are attached to it  
> with a RANDOM orientation and the polarization of illumination  
> light won't matter (no linear dichroism will be observed). But even  
> here I should stress the phrase "most cases" since there have been  
> reports on the listserver about polarization effects, especially  
> for membrane probes. Sometimes, a fluorescence image can look quite  
> different depending on the polarization of illumination. Dan  
> Axelrod has already shown how membrane blebbing or invagination can  
> be imaged using polarized TIRF illumination:
>
> Sund, S. E., J. A. Swanson, and D. Axelrod. 1999. Cell membrane  
> orientation visualized by polarized total internal reflection  
> fluorescence. Biophys. J. 77:2266-2283.
>
> John Oreopoulos
>
>
> On 24-Dec-09, at 2:12 AM, Shigeo Watanabe wrote:
>
>>
>> Dear All,
>>
>> I have previouly asked how polarized the P- and S-polarized light  
>> are at TIRF illumination.
>> Thanks to John, now I get to know that P-polarized light is  
>> converted to "cartwheel" polarized light while S-polized light is  
>> intact.
>> As he suggested I read the review of Dan Axelrod about the effect
>> (or problem) of this cartwheel polized light when observing the  
>> sample.
>> What I understand in his review is that cartwheel light excites  
>> only molecules which is parallel to z-axis.
>>
>> Researchers who I talked with about TIRF problem also mentioned  
>> that normal TIRF system which use single incident light from one  
>> entering direction excite only a fraction of  molecules which is  
>> paralle to poloized evanescent light direction and then they  
>> prefer to use the ring-like illumination for TIRF to excite every  
>> single molecules.
>>
>> Now I am confused about the actual TIRF problem.
>> Quesitons I have are these.
>> 1)Does evanescent light excite the only molecules which are  
>> paralle to polarization of evanescent light because of cartwheel  
>> polarized light even when incident light is non-polarized light???
>> I am wondering what the actual polarization of evanescent light  
>> produced by non-polarized incident light, which is mixture of P-  
>> and S- polarized light.
>>
>> 2)Do commercial TIRF systems have this polarized problem? Do they  
>> use single incident light or ring-like incident light?
>>
>>
>> I appreciate if anyone help to answer these questions.
>>
>> Sincerely
>> Shigeo Watanabe
>> Hamamatsu Photonics KK
>>
>>
>>
>>
>>
>>
>>
>>
>>