2) Design of the electron microscope laboratory (Practical methods in electron microscopy) by Ron Alderson - dated 1975 - ASIN: B0007AI2K8
3) Book chapter 4 in Imaging Cellular and Molecular Biological Functions (Principles and Practice) by Shorte & Frischknecht ISBN 9783540713302
4)Article in Wiley's Encylopeadia of Life Science PDF Light Microscopy Imaging Facilities a0022189.
> From: Nessler, Randy A <
[hidden email]>
> Subject: Re: book on "designing an Imaging Facility"
> To:
[hidden email]
> Date: Wednesday, 20 January, 2010, 19:48
> We just got done with room
> renovations for an electron microscope where the
> manufacturer stated that temperature must be stable within
> 0.5C/hr. Thus, I think you might want to ask the different
> vendors what their room requirements in regards to
> temperature, fields, and vibrations are so that you won't
> jeopardize meeting manufacturers' resolution specs.
>
> Most EM manufacturers have a pdf with these specs (minimum
> lab door sizes usually included).
>
> Randy
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:
[hidden email]]
> On Behalf Of Craig Brideau
> Sent: Wednesday, January 20, 2010 1:40 PM
> To:
[hidden email]
> Subject: Re: book on "designing an Imaging Facility"
>
> Four critical factors to consider:
>
> Temperature Control - Maintain temperature to within 2
> degrees C
> Humidity Control - Maintain humidity between 20-40%
> Vibration Control - Sufficient anti-vibration tables or
> platforms
> Electrical Supply - Have circuits with the correct voltage
> and current
> handling capacity available
>
> Also, delivery access is important. How easy will it
> be to get large
> pieces of equipment into the facility? Are there
> large cargo
> elevators available?
>
> Craig
>
>
> On Wed, Jan 20, 2010 at 12:32 PM, andre wong
> <andywong@interchange..ubc.ca>
> wrote:
> > Hi, all,
> >
> > I need a book or some informations on "how to design
> an Imaging Facilty"
> > that will accomodate a few electron microscopes, a
> couple of confocals, a
> > few other microscopes and some other ancillary
> equipment. Any information on
> > the subect will be much appreciated.
> >
> > Thanks
> >
> > Kung
> >
>