Is this the diametre of your deconvolved bead? If it is not than you have diffraction to thank for the increased width. The .17 micron beads should be sub-resolution on your microscope and are representative of a point source of light. The resulting "blurred" image that you have captured is the result of the physical properties of light and the optical constraints of your microscope. A good source for further information can be found at the following links:
But also try the Handbook for Biological Confocal Microscopy, James Pawley is the editor, for further information.
Best,
Fred
On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote:
Hey Fellow ListServers,
I purchased a PSF kit from Invitrogen and imaged the beads and when i
loaded the beads to correct the Z blur i measured the XY of the bead
to make sure the diameter matched but it turns out it doesnt. The
beads are supposedly .17um with +-.005 but whatever angle i measure
the bead at in xy i keep getting .4 um. Any ideas as to why this is
happening?
Thank You
-------
Matiar Jafari