http://confocal-microscopy-list.275.s1.nabble.com/Fluoview-1000-and-FCS-tp4594644p4621592.html
>"time mode" which counts photons/time-block, or "photon mode" which counts each
>photon with its respective time).
>
>In ISS Vista, we always exclude the first data point in the analysis, because
>it
>has no preceeding data points to be correlated with for the correlation
>analysis.
>I cant recall excluding the first point when we have analysed FV1000 point scan
>data in SimFCS, I'm not sure if there is an option to do so. If the first data
>point can be excluded in the analysis I would do so. Do not simply delete the
>first data point, as this would simply mean that now the second data point has
>no first point to be correlated to so the problem still remains.
>If the first point can not be excluded from the analysis, it is only one point
>whose correlation will be of reduced accuracy, out of the 32766 points, so I
>dont think this matters very much, and perhaps (? - maybe a mathematician can
>answer) it is not as important in time mode as in photon mode.
>
>SimFCS is definately worth trying, yes 30day demo licence is free, and the full
>licence is very economical anyway. Go to the website for the Laboratory for
>Fluorescence Dynamics;
http://www.lfd.uci.edu/>
>For point-scanning data collection on the FV1000, configure for optimal
>resoluion, ie use a high NA objective etc, select photon counting mode, ie
>"photon cnt", chose appropriate pixel time eg 8-12.5us/pixel for solution or
>12.5-40us/pixel for measurement on cells, select crosshair icon and set desired
>point in image, set number of frames to 32766 (if you type in an overly high
>number eg 1000000 it will automatically reset to the maximum available) with
>interval set to "freerun". Make sure you export the data as the "raw" 16-bit
>files.
>
>If you need a hand working through the analysis in the SimFCS software contact
>me offline.
>Note that the Wo (PSF waist) in SimFCS is defined using 1/e2 max intensity not
>half max intensity.
>
>Its always a good idea to include a reference dye for which you already have a
>good idea of expected or theoretical diffusion coefficient.
>
>Hope this helps
>Kind regards
>Jennifer
>
>--
>Jennifer Clarke BSc (Hons) PhD
>Research Associate, Anatomy and Histology
>Centre for Neuroscience, School of Medicine
>&
>Facility Manager, Optical Microscopy Suite, Flinders Microscopy
>
>Flinders University
>GPO Box 2100, Adelaide 5001
>Phone: 61 8 8204 6637 / 61 8 8204 6454
>Email:
[hidden email]
>
>--------------------------------------------------------
>
>Quoting Stanislav Vitha <
[hidden email]>:
>
>> Hi,
>> I second Aseem's suggestion regarding SImFCS.
>>
>> A user of our core facility is using FV1000 and the SimFCS software from the
>> Gratton lab, it appears to work quite well, as far as I know. The SimFCS
>> software can be installed as a fully-functional 1-month demo, so you can
>> test it yourself before making the final decision. One big advantage of the
>> RICS approach that I see is the ability to subtract slow movements, such as
>> sample or stage drift.
>> A while ago I was told by Olympus that point scanning was not going to work
>> for FCS on the FV1000 system, but I do not quite remember their technical
>> explanation. From my tests, I remember there was an issue with the first or
>> last datapoint being of incorrect value, due to the timing of the AOTF
>> "shutter", or synchronization of the AOTF with data acquisition. You could
>> try point-scanning of the fluorescent plastic slide from Chroma, the
>> fluctuations that you see should indicate the instrument noise/instability,
>> including the laser.
>> The number of datapoints in point-scanning mode on FV1000 was limited to
>> 4096, i.e., the max. permitted line width. I was told that before a scan a
>> piece of code is sent to the control board(s), essentially reprogramming the
>> firmware, and this somehow imposes a limit on the number of datapoints.
>> You could do time-lapse point scan to acquire multiples of 4096 points,, but
>> then you have an unknown amount of time between the time frames.
>>
>> Anyway, I think SimFCS is worth trying.
>>
>> Stan Vitha
>> Microscopy and Imaging Center
>> Texas A&M University
>>
>> On Sat, 20 Feb 2010 22:09:52 +0530, aseem mishra <
[hidden email]>
>> wrote:
>>
>> >Hi Jean,
>> >
>> >You might want to have a look at Enrico Gratton's work and specially
>> >SimFCS wherein one can calculate diffusion coefficients from an image
>> >collected in a point-scan mode. I do not know if a blinking laser
>> >gives you the possibility of doing FCS (Do let me know if such a
>> >possibility exists. I would love to try it at my place.) We rather use
>> >a hardware correlator and a simple laser (diode/HeNe)
>> >
>> >Aseem
>> >
>> >
>> >On Fri, Feb 19, 2010 at 3:27 AM, Jean-Pierre CLAMME
>> ><
[hidden email]> wrote:
>> >>
>> >> Hi all,
>> >>
>> >> I'm thinking about modifying our fluoview 1000 to do FCS. The fluoview
>> has
>> >> the possibility to do single point measurement. However as we can still
>> >> choose the pixel time in this mode, I was wondering if the laser beam
>> always
>> >> stays at the same point of interest and continuously illuminates the
>> point
>> >> or if it actually "blinks".
>> >>
>> >> If it doesn't "blink", I would think it could hinder the FCS measurement
>> et
>> >> thus I'm wondering if there is an option to change this in the soft. ?
>> >>
>> >> Thanks,
>> >>
>> >> JP
>> >>
>> >>
>> >>
>> >>
>> >> - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
>> >> Jean-Pierre CLAMME, PhD
>> >> Senior Scientist
>> >> Nitto Denko Technical
>> >> 501 Via Del Monte
>> >> Oceanside, CA 92058
>> >> E-mail:
[hidden email]
>> >> Phone: +760.435.7065
>> >>
>> >
>> >
>> >
>> >--
>> >Aseem Mishra
>> >Senior Research Assistant,
>> >Malaria Group,
>> >International Centre for Genetic Engineering and Biotechnology,
>> >New Delhi-110067
>> >INDIA
>>
>
>
>--
>Jennifer Clarke BSc (Hons) PhD
>Research Associate, Anatomy and Histology
>Centre for Neuroscience, School of Medicine
>&
>Facility Manager, Optical Microscopy Suite, Flinders Microscopy
>(available for training and assistance on Mondays only)
>
>Flinders University
>GPO Box 2100, Adelaide 5001
>Phone: 61 8 8204 6637 / 61 8 8204 6454
>Email:
[hidden email]