Re: tube lens and spherical aberration
Posted by
mmodel on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Southeastern-Microscopy-Society-meeting-tp4640594p4642705.html
Andreas, thanks for your response. I understand that the path can be reversed for a ray but it seems to me that the size of the blur in the image will depend, for example, on the distance between the
objective and the tube lens. Which is not part of the standard formula.
Mike
Just go the other way, from a perfect spot in the image plane to the diffuse spot in the object plane, then you are back to the standard situation. You can always reverse the light path in linear
optics.
best wishes
Andreas
-----Original Message-----
From: MODEL, MICHAEL <
[hidden email]>
To:
[hidden email]
Sent: Fri, 26 Feb 2010 20:54
Subject: tube lens and spherical aberration
(I apologize if this is a second post, I got some strange errors from the first
one)
Dear List
It's not a confocal question, it's strictly a wide-field question... When people
talk about spherical aberration they typically start with a parallel beam
falling on an objective, and then, for whatever reason, the objective fails to
bring all the rays into a common focus. In wide-field, one has a luminous spot
in the object space, and the effect of spherical aberration would be a failure
of the objective to collect all the rays into a single parallel beam. Some rays
will form a converging (or diverging) cone instead. Then, this cone will be
received (I suspect that in the case of a diverging cone, some light may even
get lost on its way) by a tube lens and form a blurry spot on the image plane.
My question is, Is this situation really equivalent to the standard one
considered in all books? (I suspect, it's not because even the distance to the
tube lens should make a difference). But I would be very interested in the
opinion of those who understand optics better than I do. Thanks!
Mike Model