Just to mention, should one be stuck with PMTs instead GaAs,   one could play with the applied bias voltage to modify   dark noise   (to  

  balance the  gain   versus noise.  )

Nice thing with Olympus Kalman filtering is   that its use would allow   increase the bias voltage of PMT   

 

 

 

 

Thanks

 

Axel   Central Imaging  (IMCB)  6-19B,  cell  +65 9271.5622

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
Sent: Wednesday, March 10, 2010 8:17 AM
To: [hidden email]
Subject: Re: Zeiss or Olympus

 

Mark
 
I was thinking about the use of GaAs ( gallium asrenide) detectors for multiphoton by Zeiss for the NLO.  This are not a PMT and have very different properties. 
 
Dick Burry

----- Original Message -----
From: Mark Cannell <[hidden email]>
Date: Tuesday, March 9, 2010 6:33 pm
Subject: Re: Zeiss or Olympus
To: [hidden email]

> Hi RICHARD
>
> As far as I know, for PMT based systems the dynamic range does
> not
> depend on PMT sensitivity (which is determined by the
> photocathode) but
> on gain (=PMT HV) and A/D settings.
>
> Cheers Mark
>
> BURRY wrote:
> >
> > Vladimir and Stephan
> >
> > 
> >
> > The Zeiss 780 is an upgrade for the single photon.  The
> big issues for
> > me with the Zeiss and Olympus multiphoton are the detector
> dynamic
> > range, and detector sensitivity.  Extremely sensitive
> detectors can
> > have a limited dynamic range.  I would be interested in
> hearing from
> > someone who had recently tested not only the sensitivity of
> these two
> > multiphotons, but also the dynamic range in an image collected
> a
> > moderate to low fluorescence levels.  My evaluation a
> year ago was
> > that the Olympus multiphoton had the better combination of
> sensitivity
> > and dynamic range.
> >
> > 
> >
> > Dick Burry
> >
> >
> > ----- Original Message -----
> > From: Stephan Speier <[hidden email]>
> > Date: Tuesday, March 9, 2010 8:10 am
> > Subject: Re: Zeiss or Olympus
> > To: [hidden email]
> >
> > > Dear Vladimir,
> > >
> > > We tested the Zeiss LSM710 and the Olympus FV1000 regarding
> > > confocal
> > > and multiphoton imaging and had already decided for the FV1000
> > > when
> > > Zeiss released the LSM780.
> > > After testing the LSM780 we changed our mind and decided for
> Zeiss.> > For our purposes the LSM780 has exactly the
> improvements over
> > > the 710,
> > > we were looking for and therefore beat the Olympus.
> > >
> > > Best,
> > > Stephan
> > >
> > >
> > > ----- Original Message -----
> > > From: Vladimir Gukassyan <[hidden email]>
> > > Date: Tuesday, March 9, 2010 1:18 pm
> > > Subject: Zeiss or Olympus
> > > To: [hidden email]
> > >
> > > > Dear List Members,
> > > >
> > > > We're making a selection between Zeiss 710 and Olympus FV1000
> > > for the
> > > > multiphoton imaging.
> > > > I would be thankful for all user experience cases and
> > > suggestions on
> > > > this matter.
> > > >
> > > > Thank you in advance.
> > > > With regards,
> > > > Vladimir
> > > >
> > > > ---------------------------------
> > > > Vladimir Ghukasyan
> > > > Confocal and Multiphoton Imaging Core,
> > > > Neuroscience Research Center
> > > > University of North Carolina
> > > > 115 Mason Farm Rd., Bld. 245, 7 Fl.
> > > > Chapel Hill, NC 27599-7250
> > > > Tel.: +1 919 966 5807
> > > >
> > >
> > >
> > > --
> > >
>
>
> --
> BEGIN-ANTISPAM-VOTING-LINKS
> ------------------------------------------------------
>
> Teach CanIt if this mail (ID 1010662020) is spam:
> Spam:       
> https://antispam.osu.edu/b.php?i=1010662020&m=f29e0aefb64a&c=sNot spam:    https://antispam.osu.edu/b.php?i=1010662020&m=f29e0aefb64a&c=n
> Forget vote:
> https://antispam.osu.edu/b.php?i=1010662020&m=f29e0aefb64a&c=f---
> ---------------------------------------------------
> END-ANTISPAM-VOTING-LINKS
>

Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849


Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.