Dear listserv,
GaAsP vs conventional PMT quantum efficiency curve is on the bottom of
page 19 of the Zeiss LSM 780
Product Brochure (7.3 MB) 60-1-0015_lsm_780_e[1].pdf at
http://www.zeiss.de/C12567BE0045ACF1/Contents-Frame/0AE7CE3138743B6CC12576400050DF64
The 2010 Zeiss On Your Campus live cell imaging workshop CD (DVD?)
"Single Molecules: Going Beyond Basic Imaging" pdf, page 16 (of
58) has silicon photodiode (APD), Ga-As-P PMT and Ga-As PMT QE curves
(page 15 has CCD curves). If you are not attending/hosting a ZOYC, ask
your local Zeiss rep for a copy of the CD or talk pdf's. The graphs in
the ZOYC pdf look like molecular expressions figures, so you might also
find them at M.E. or the M.E. hosted Zeiss web site.
As for Jim's mention that the GaAsP and conventional PMT cross-over at
around 700 nm (maybe 670 nm in ZOYC graph): in almost 3 years at UM I
cannot recall any user bringing a specimen with a fluorophore whose
emission maximum longer than 700 nm. That said, a new paper, Tonninges et
al 2010 J Microscopy has several new PTIR and NeuroVue lipophilic
dyes with nice 1p and 2p excitation and NIR emission properties. Paper is
on neurotracing, but may also be useful as a DiI replacement for blood
vessel painting if the price is right.
Sincerely,
George
At 04:28 PM 3/11/2010, you wrote:
From the Zeiss
Brochure:
The new GaAsP detector technology is not confined to visual
light excitation. The LSM BiG (binary GaAsP) now also offers
you multicolor multiphoton imaging with GaAsP performance.
Two LSM BiG modules can be added to NLO systems
as transmitted and incident light NDDs, providing 4 ultrasensitive
detection channels. The LSM 780 NLO and LSM 710 NLO
let you penetrate deeper and detect more light.
So additional to the internal GaAsP spectral detector you can add a GaAsP
NDD (BIG detector) for multiphoton imaging. And you can have the GaAsP
single channel sitting directly over the objective. I think you can also
couple two GaAsP detectors on an external port of the scan head
best wishes
Andreas
Thanks Andreas,
Does anyone have any idea of the maker or model number of these BiG
detectors?
As I mentioned earlier, the count-rate performance of APD single-photon
counting systems are limited by the capacitance of the sensor diode. This
is in turn proportional to the the sensitive area of the
detector.
For this reason, the sensitive areas tend to be small: maybe 100µm
diameter. As this is much smaller than the ray bundle coming out the back
of most high-mag, high-NA objectives, it is not obvious how one could
"squeeze" the light bundle there to match the sensitive area.
So I assume that the word "big" in your post is relevant and it
would be nice to know how big and if it works in the counting mode, and
whether it is an avalanche diode or merely a GaAsP photodiode (no
amplification, but no multiplicative noise either. Very suitable for
transmitted detectors....).
Cheers,
Jim P.
**********************************************
Prof. James B.
Pawley,
Ph. 608-263-3147
Room 223, Zoology Research
Building,
FAX 608-265-5315
1117 Johnson Ave., Madison, WI, 53706
[hidden email]
3D Microscopy of Living Cells Course, June 12-24, 2010, UBC, Vancouver
Canada
Info:
http://www.3dcourse.ubc.ca/
Applications due by March 15, 2010
"If it ain't diffraction, it
must be statistics." Anon.
-----Original Message-----
From: RICHARD BURRY <[hidden email]>
To: [hidden email]
Sent: Wed, 10 Mar 2010 18:54
Subject: Re: Zeiss or Olympus
Jim
Thanks for the details of the GaAs and GaAsP detectors. But my
question remains, does Zeiss use either of these detectors in its 780
multiphoton? And does this represent an improvement in S/N over
standard PMTs? My experience is with the NLO a while ago is that
the GaAs detector was more sensitive but for samples with a range of
emission intensity, it was difficult to not saturate some part of the
sample.
Dick
----- Original Message -----
From: James Pawley
<[hidden email]>
Date: Wednesday, March 10, 2010 11:58 am
Subject: Re: Zeiss or Olympus
To:
[hidden email]
> Just to clarify, the 780
has a GaAsP (Gallium Arsenite Phosphate) detector, not GaAs, the
difference in quantum efficiency can be seen e.g. in the Webb multiphoton
review (Nature Biotechnology 2003, 21, 1369). The drawback is that GaAsP
QE drops dramatically for wavelength > 700 nm, but they put a normal
PMTs as the two additional channels on the 780, to cover the extended
range. By the way GaAsP detectors are PMTs as well, it is just a
different material of the photocathode, afterwards the photoelectrons are
multiplied in the same way. GaAsP detectors reach 40% quantum efficiency
which is about twice as sensitive as a normal PMT. APDs have 60-70% and a
back-thinned CCD about 90%., so still a lot of signal is thrown away, not
to mention the losses on the way to the detector.
> Andreas
> Hi all,
> Indeed, the GaAs and GaAs phosphide QE curves are very impressive.
However, it is important to remember what is actually measured to make
these curves. PMT curves refer to the fraction of photons striking the
photocathode that produce photoelectrons (It is usually measured by
allowing a calibrated amount of light to strike the photocathode and
using a nano-ammeter to sense the total photoelectron current between the
photocathode and all the other electrodes in the PMT). However, depending
on the electrode geometry, 10-30% of these photoelectrons don't actually
hit the first dynode (D1), and therefore do not contribute to the PMT
output.
> Of those photoelectrons that do hit D1, a reasonable fraction fail
to excite any secondary electrons, and again, do not contribute to the
PMT output. There are many reasons for this but one is just Poisson
statistics. If the average gain is say 4, then about 8% of the collisions
will result in zero electrons being emitted. However, this effect is
again multiplied by geometrical factors where SE produced in different
parts of D1 have better or worse chances of actually striking D2 and
producing a SE. Signal loss in this way depends a lot on the actual
voltage between the photocathode and D1: it will be less when the voltage
is higher. Unfortunately, few confocals seem to have been set up in
such way that this is always true. On average signal loss by
failure to propagate after collision with D1 will be an additional
20-40%.
> Finally, the same type of Poisson effects that cause some signal to
be lost entirely, cause the amount by which the remainder is amplified to
be highly variable (10-90%). This variation degrades the accuracy of the
output signal by introducing what is called multiplicative noise. Because
this extra noise can only be "overcome" by counting more
photons, its presence effectively reduces the effective QE of the
device. In the best case, this reduction is about 40% and in the worst
case (an exponential gain distribution, approximated by some micro PMTs)
75% (i.e., the QE is reduced to 60% or 25% of what it would have been if
all photoelectrons were equally amplified).
> As a result, while the peak effective QE of a PMT with a GaAs
or GaAsP photocathode is indeed much better than that of the more common
S-20 photocathode, in terms of its effectiveness in providing an output
current that is proportional to the input photon signal, the QE is more
in the range of 3 -10% (depending on dynode geometry and first-dynode
voltage) than 40%. (The 60% numbers are for APDs rather than for a GaAs
or GaAsP photocathode on a PMT.)
> The performance can be improved somewhat on the few confocals that
allow single-photon counting as this procedure eliminates multiplicative
noise. (see below about the limitations imposed by photon
counting)
> This tedious recital is I hope justified by noting that, at
least when EG&G was the major APD supplier, APD performance was not
specified in terms of QE but as Photon Detection Efficiency (PDE).
Although APDs can be operated in a low gain, proportional mode, their PDE
under these conditions is very low (because APD multiplicative noise is
very high and at low (non-avalanche breakdown) gain, by far the most
likely gain of the initial photoelectron is zero).
> Therefore, high PDE (or high QE) AOD units tend to operate at high
bias (high, avalanche gain) and this requires circuitry to quench the
avalanche breakdown and count the single-photon pulses. Modern units
contain both the sensor itself and the pulse counting and avalanche
quenching circuits needed for counting the single-photon pulses. In other
words (assuming that Hamamatsu follows the EG&G precedent), their QE
figures for single-photon counting units already include any losses for
non-propagation or multiplicative noise. Therefore, a quoted PID of 60%
really does mean that 60% of the photons (of the specified wavelength)
that strike the center of the active surface will be accurately
counted.
> This is about 4-10x better than the performance of a similar GaAs or
GaAsP photocathode on a PMT.
> This good news is tempered by the fact that, because of the high
capacitance of the AOD itself, it is hard to count much faster than, say
30MHz. As 30MHz comes out to an absolute maximum of 60 counts during a 2
µs pixel, this means that at least 50% of your counts will be lost due to
pulse pileup when 30 counts arrive per pixel and 10% will be lost at only
6 counts/pixel. In other words one has to be very careful to adjust the
excitation intensity so as not to "clip" the brightness of
those parts of the image that contain a lot of fluorophor.
> Lots more on this in The Handbook,
> Cheers,
> Jim Pawley
>
**********************************************
> Prof. James B.
Pawley,
Ph. 608-263-3147
> Room 223, Zoology Research
Building,
FAX 608-265-5315
> 1117 Johnson Ave., Madison, WI,
53706
[hidden email]
> 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,
Vancouver Canada
> Info:
http://www.3dcourse.ubc.ca/
Applications due by March 15, 2010
>
"If it ain't diffraction, it must be statistics." Anon.
> Just to mention, should one be stuck with PMTs instead
GaAs, one could play with the applied bias voltage to
modify dark noise (to
balance the gain versus
noise. )
Nice thing with Olympus Kalman filtering is that
its use would allow increase the bias voltage of PMT
Thanks
Axel
Central Imaging (IMCB)
6-19B, cell +65 9271.5622
From: Confocal Microscopy List
[[hidden email]] On Behalf Of RICHARD
BURRY
> Sent: Wednesday, March 10, 2010 8:17 AM
> To: [hidden email]
> Subject: Re: Zeiss or Olympus
Mark
> I was thinking about the use of GaAs ( gallium asrenide) detectors
for multiphoton by Zeiss for the NLO. This are not a PMT and have
very different properties.
> Dick Burry
> Note: This message may contain
confidential information. If this Email/Fax has been sent to you by
mistake, please notify the sender and delete it immediately. Thank
you.
> -- >
**********************************************
> Prof. James B.
Pawley,
Ph. 608-263-3147
> Room 223, Zoology Research
Building,
FAX 608-265-5315
> 1117 Johnson Ave., Madison, WI,
53706
[hidden email]
> 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC,
Vancouver Canada
> Info:
http://www.3dcourse.ubc.ca/
Applications due by March 15, 2010
>
"If it ain't diffraction, it must be statistics." Anon.
> Spam
> Not spam
> Forget previous vote
Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814 Cell 614.638.3345 Fax
614.247.8849
--
**********************************************
Prof. James B.
Pawley,
Ph. 608-263-3147
Room 223, Zoology Research
Building,
FAX 608-265-5315
1117 Johnson Ave., Madison, WI, 53706
[hidden email]
3D Microscopy of Living Cells Course, June 12-24, 2010, UBC, Vancouver
Canada
Info:
http://www.3dcourse.ubc.ca/
Applications due by March 15, 2010
"If it ain't diffraction, it
must be statistics." Anon.
George McNamara, Ph.D.
Image Core Manager
Analytical Imaging Core Facility
University of Miami, Miller School of Medicine
Miami, FL 33136
[hidden email]
[hidden email]
305-243-8436 office
http://www.sylvester.org/AICF (Analytical Imaging Core Facility)
http://www.sylvester.org/AICF/pubspectra.zip (the entire 2000+
spectra .xlsx file is in the zip file)
http://home.earthlink.net/~geomcnamara