There's no reason why you can't do this, and the images will be fine,
but if you're looking to do 3D sectional imaging of the RBCs, you
might be better off with a high NA water immersion objective instead
which will have lower spherical aberrations - you likely will get
better signal with that type objective. Try comparing both types of
objectives if possible.
John Oreopoulos
On 22-Mar-10, at 5:29 AM, Charu Tanwar wrote:
> Dear List
>
> Anybody please let me know whether we can use TIRF 100XH (N.A.
> 1.49,Working Distance 0.12mm,Coverglass correction 0.13-.20mm -
> this is a
> special TIRF objective) for routine confocal imaging of bacteria
> and RBC's with
> low flourescence signal.
> Can it be mounted on a confocal microscope from TIRF microscope if
> both the
> systems are from same company.
>
> Thanks in advance
>
> Charu Tanwar
> Imaging Specialist
> Advanced Instrumentation Research Facility
> Jawaharlal Nehru University
> New Delhi
> India.