Re: TIRF objective for routine imaging

Posted by Peter Gabriel Pitrone on
URL: http://confocal-microscopy-list.275.s1.nabble.com/TIRF-objective-for-routine-imaging-tp4778151p4784145.html

Hello Charu,

If using this objective on a CLSM is your plan, then I would go for a lower magnification (the 60x/1.49 instead of the 100x/1.49 for example) due to transmission efficiencies. Objective with the same N.A. but with lower magnifications are almost always a better choice when working with point scanners. Just "zoom" in to meet your Nyquist!

Pete

On Mar 23, 2010, at 04:55 AM, charu tanwar wrote:

Thank you all for your valuable inputs.
Charu

CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.


--- On Tue, 23/3/10, Tim Feinstein <[hidden email]> wrote:

From: Tim Feinstein <[hidden email]>
Subject: Re: TIRF objective for routine imaging
To: [hidden email]
Date: Tuesday, 23 March, 2010, 5:55 AM

We use the same objective as Neeraj on a A1/TiE for confocal, widefield and TIRF.  No complaints. 

cheers, 


Tim Feinstein


On Mar 22, 2010, at 7:39 PM, Neeraj Gohad wrote:

Hi Charu,
 
We have A Nikon TIRF module on our Nikon TiE so we have the Apo 60X 1.49 NA TIRF Objective, I have used this for regular confocal imaging with great success.
 
Best,
 
Neeraj.
 
Neeraj V. Gohad, Ph.D.
Postdoctoral Fellow
Okeanos Research Group
Department of Biological Sciences
132 Long Hall
Clemson University
Clemson,SC-29634
Phone: 864-656-3597
Fax: 864-656-0435
 
 
 
 
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Marco Dal Maschio
Sent: Monday, March 22, 2010 3:56 PM
To: http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@...
Subject: Re: TIRF objective for routine imaging
 
Dear Charu,
I had the opportunity to test from the same company 60x 1.45tirf // 60x VC // 60x 1.49 tirf.
I remember that probably 60x 1.49 is not plan, related to the  flat field correction. 
But this objective is currently used by colleagues  performing tracking and imaging in neuronal cultures
with optimal image quality. If you want I can send you images comparing 60x VC and 60x 1.49 tirf.
Sorry but no experience about 100x.
 
best 
Marco   
 
On Mon, Mar 22, 2010 at 10:29 AM, Charu Tanwar <http://in.mc83.mail.yahoo.com/mc/compose?to=tanwar_charu@...> wrote:
Dear List

Anybody please let me know whether we can use TIRF 100XH (N.A.
1.49,Working Distance 0.12mm,Coverglass correction 0.13-.20mm - this is a
special TIRF objective) for routine confocal imaging of bacteria and RBC's with
low flourescence signal.
Can it be mounted on a confocal microscope from TIRF microscope if both the
systems are from same company.

Thanks in advance

Charu Tanwar
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India.





        




Your Mail works best with the New Yahoo Optimized IE8. Get it NOW!.