Hello Charu,
If using this objective on a CLSM is your
plan, then I would go for a lower magnification (the 60x/1.49 instead
of the 100x/1.49 for example) due to transmission efficiencies.
Objective with the same N.A. but with lower magnifications are almost
always a better choice when working with point scanners. Just
"zoom" in to meet your Nyquist!
Pete
Hi Charu and Pete,
"Efficiency" is a tricky thing. Everyone is in favor of
it but it often means different things to different people.
In the days of widefield systems with Hg arcs and excitation
optics that made a large image of the arc in the objective BFP, there
was a rumor that the "brightness" of a fluorescent image
would vary with the fourth power of the NA and inversely with the
square of the magnification. So high NA, 40x (and now 20x!) objectives
became popular. The problem was that this brightness was related to
visibility by eye. Once you wanted to record the low-mag image on
film, you had to use slower film with finer grain to preserve the
details (Ok. Many people didn't care about the details and used fast
film just to see if there was some stain, but this approach lost the
details.)
When we got CCD cameras, we all became a bit more conscious of
Nyquist sampling and realized that lower mag meant smaller camera
pixels and, in the end, we collected the same number of photons/pixel
no matter what objective mag you choose. Assuming that you had
settled on a camera built around one of the ubiquitous SONY chips
designed for the Japanese HDTV standard, the pixels were 6.7µm and
you would just have to choose the right coupling tube mag to allow you
to meet Nyquist.
So that is the mag part: In terms of image intensity, mag isn't
important with laser instruments (including laser TIRF), unless you
need low lag to cover a larger field and are willing to scan slowly,
with a huge raster and tolerate some off-axis image degradation.
What about the NA part? As a mercury arc is about the same
dimensions as the BFP of the objective (say 3-5 mm. It varies with arc
power: 50w, 100w, 200w), any efficient illumination system will focus
this into the BFP at about 1:1 because 1:1 optics have the highest
potential light throughput.
Therefore, the objective that accepts light from the largest
fraction of this arc-image and conveys it to the specimen will excite
the most fluorescence. As the diameter of the BFP varies directly with
NA, its area varies with (NA)*2. As the same should follow for the
efficiency with which the objective collects the excited fluorescence,
it will also vary with (NA)*2 and the total overall brightness will
vary with (NA)*4.
Some BFP background
All other things being equal, a 100x objective has a focal length
that is 2.5x shorter than a 40x lens. If, for a moment, you can just
imagine the simplest lens diagram with on-axis, parallel illumination
striking the lens and then converging to form a focused spot, one
focal length away from the optical center of the lens (the optical
center is the plane about which it seems to act). No matter what the
objective magnification, the boundary of the angle of convergence is
set only by the NA. Therefore, as the 40x lens has a 2.5x longer focal
length, its principle plane must be 2.5x farther from the focal plane
compared to that of the 100x lens. And as the divergence is the same,
from similar triangles, the diameter of the parallel ray bundled
needed to fill the 40x lens must be 2.5x larger than that needed to
fill the 100x lens.
Put simply, the diameter of the clear aperture that you see when
you look at a 100x NA 1.3 is about only about 40% as large as that
seen when looking at the back of a 40x NA 1.3, and if you focus a
more-or-less uniform image of the arc onto this plane, and it is big
enough to "fill" the BFP of the 40x, then you will find that
about (2.5)*2 = 6.25x more light comes out of the 40x, compared to the
100x. (On the other hand, if your excitation optics make an image of
the arc in the BFP that is much smaller than that of the 40x BFP, this
effect will be proportionally less. Assuming Kohler illumination, you
should be able to see the relative size of the arc and the BFP if you
just let the illumination proceed from the objective and fall onto a
piece of lens tissue placed a cm or two on the far side of the focus
plane.)
However, if you have a laser-based illumination (such as is found
on many laser confocals), things are a little different. First all
all, on the illumination side, excitation intensity itself is not a
usually a problem. In fact we have to adjust the laser power down to
prevent singlet-state saturation and undue photodamage. So one of the
two (NA)*2 intensity terms become irrelevant. (On the imaging side, of
course, larger NA will collect more light and, in the absence of
spherical aberration etc, will deliver more photons to the
detector.)
On the other hand, in laser-based systems, the diameter of the
diffraction-limited spot is inversely proportional not to the NA
engraved on the objective barrel but to the fraction of this NA
actually filled by the laser beam. A 1mm diam unexpanded laser beam
filling a 60x 1.3NA lens, with a BFP about 8mm in diameter will form a
focused spot that is 8x LARGER than it should be. (The confocal image
may still look sort of OK as the NA on the imaging side will
still be 1.3, but the X,Y and Z resolution will still be lower.)
Now, confocal manufacturers know all this and they try to set
things up so that the laser beam is large enough to "fill the
BFP" of the most common lenses with more-or-less uniform laser
intensity. However, you can see their problem, unless the illumination
optics actually adjusts to match each lens, then we are back to the
widefield system: if it is adjusted so that the laser "fills"
the 40x BFP, then only 1/6th of that light will be used with the 100x.
The rest will be bouncing around causing troublesome reflections. So
they compromise, with the result that, in practice, one will often
record slightly better resolution in X, Y and Z with the higher
mag objective because the illumination will better fill its
aperture.
So wasn't that a long story. I hope you are not all asleep (but I
will be checking!)
Sorry. Must be Spring! And sorry for the typos...
Jim P.
**********************************************
Prof. James B.
Pawley,
Ph. 608-263-3147
Room 223, Zoology Research
Building,
FAX 608-265-5315
1117 Johnson Ave., Madison, WI, 53706
[hidden email]
3D Microscopy of Living Cells Course, June 12-24, 2010, UBC, Vancouver
Canada
Info:
http://www.3dcourse.ubc.ca/
Applications
still being accepted
"If it ain't
diffraction, it must be statistics." Anon.
On Mar 23, 2010, at 04:55 AM, charu
tanwar wrote:
Thank you all for your valuable
inputs.
Charu
CHARU TANWAR
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi 110067
India.
--- On Tue, 23/3/10, Tim Feinstein <[hidden email]> wrote:
From: Tim Feinstein <[hidden email]>
Subject: Re: TIRF objective for routine imaging
To: [hidden email]
Date: Tuesday, 23 March, 2010, 5:55 AM
We use the same objective as Neeraj on a A1/TiE for
confocal, widefield and TIRF. No complaints.
cheers,
Tim Feinstein
On Mar 22, 2010, at 7:39 PM, Neeraj Gohad
wrote:
Hi Charu,
We have A Nikon TIRF module on our Nikon TiE so we have
the Apo 60X 1.49 NA TIRF Objective, I have used this for regular
confocal imaging with great success.
Best,
Neeraj.
Neeraj V. Gohad, Ph.D.
Postdoctoral Fellow
Okeanos Research Group
Department of Biological Sciences
132 Long Hall
Clemson University
Clemson,SC-29634
Phone: 864-656-3597
Fax: 864-656-0435
Website: http://www.clemson.edu/okeanos
Please note my new email address: http://in.mc83.mail.yahoo.com/mc/compose?to=neerajg@...
From: Confocal Microscopy List
[mailto:[hidden email]] On Behalf
Of Marco Dal Maschio
Sent: Monday, March 22, 2010 3:56 PM
To: http://in.mc83.mail.yahoo.com/mc/compose?to=CONFOCALMICROSCOPY@LISTS.UMN.EDU
Subject: Re: TIRF objective for routine
imaging
Dear Charu,
I had the opportunity to test from the same company 60x
1.45tirf // 60x VC // 60x 1.49 tirf.
I remember that probably 60x 1.49 is not plan, related to
the flat field correction.
But this objective is currently used by colleagues
performing tracking and imaging in neuronal
cultures
with optimal image quality. If you want I can send you
images comparing 60x VC and 60x 1.49 tirf.
Sorry but no experience about 100x.
best
Marco
On Mon, Mar 22, 2010 at 10:29 AM, Charu Tanwar <http://in.mc83.mail.yahoo.com/mc/compose?to=tanwar_charu@...> wrote:
Dear List
Anybody please let me know whether we can use TIRF 100XH (N.A.
1.49,Working Distance 0.12mm,Coverglass correction 0.13-.20mm - this
is a
special TIRF objective) for routine confocal imaging of bacteria and
RBC's with
low flourescence signal.
Can it be mounted on a confocal microscope from TIRF microscope if
both the
systems are from same company.
Thanks in advance
Charu Tanwar
Imaging Specialist
Advanced Instrumentation Research Facility
Jawaharlal Nehru University
New Delhi
India.
Your Mail works best with the New Yahoo
Optimized IE8. Get it NOW!.
--
**********************************************
Prof. James B.
Pawley,
Ph. 608-263-3147
Room 223, Zoology Research
Building,
FAX 608-265-5315
1117 Johnson Ave., Madison, WI, 53706
[hidden email]
3D Microscopy of Living Cells Course, June 12-24, 2010, UBC, Vancouver
Canada
Info:
http://www.3dcourse.ubc.ca/
Applications due by March 15,
2010
"If it ain't
diffraction, it must be statistics." Anon.