Hi Steve

As EM gain calibration is so trivial -I couldn't help but be unimpressed
:-P  To calibrate in terms of average photoelectrons across the image is
also trivial when you reduce the signal to << 1 photon per pixel and
take plenty of frames. But that is not still not accounting for the
pixel to pixel sensor variation.  My point is that  is that you can't
calibrate an _image_ by assuming that the gain and offset of every pixel
is the same -you need darks and flats to do this and only then can you
provide an image calibrated in 'photons' captured. I may be getting old
but I don't like seeing quite complicated ideas being distilled by "turn
key" solutions to the point where a user thinks they have something
accurately calibrated -but never know what the calibration means or its
assumptions.

I know that many folks these days don't seem to want to know anything
about the limitations of the methods they use because they think results
are more important (than actually understanding what their machine
actually does). But as you know, my view is that unless you "understand
the process of imaging you risk imagination" (c).

another 2c.

Cheers  Mark

Stephen Cody wrote:

> Dear List,
>
> I just checked the Evolve web page again, it is not a "Dark
> Calibration" but a light calibration. A shutter is closed, and an
> internal light source in the camera activated to calibrate the EMCCD.
> Below I've extracted the relevant text..... I have no commercial
> affiliation with Photometrics.
>
> From Photometrics Web page (This is from a commercial company, I have
> no affiliation ad I have no personal experience of this product).
>
> "EMCCD cameras are subject to aging of the EMCCD register as a result
> of its usage. The Evolve has a simple calibration feature that
> performs the industry’s most accurate EM calibration within 3 minutes.
>
> A simple turn of the camera’s nose-piece closes a shutter and
> activates a light source which the detector uses to calibrate its EM
> gain. This ensures that users will receive the most accurate EM gain
> and EM gain applied matches what the user requests.
>
> Simple software control will allow the user to use this feature as a
> manual shutter in order to block all light from the sensor in order to
> take dark reference frames if necessary."
>
> On 20 April 2010 09:37, Stephen Cody <[hidden email]> wrote:
>  
>> Dear Mark et al,
>>
>> As I understand from the promotional material for this camera, there
>> is a dark calibration procedure built into the camera. The eVolve web
>> site while very glittzy is quite informative (if you can stand the
>> hype)
>>
>> Stephen Cody
>>
>> On Saturday, April 17, 2010, Mark Cannell <[hidden email]> wrote:
>>    
>>> Hi All
>>>
>>> I must admit to being unimpressed by this 'improvement'. It removes (from the researcher) the need to understand what a camera really does and I doubt that it is accurate. Before someone howls at this, I would point out that astronomers who routinely produce calibrated images use a dark and a flat frame to achieve this. Without a dark, you cannot calibrate the camera image -even if you assume it is flat (which it isn't). The problem is that the camera changes it's properties (especially the EM register) so no single calibration is going to be accurate. Since it is easy to actually use darks and flats to calculate actual photon numbers, why rely on a manufacturer calibration? I suggest it's a bit like assuming your Gilson/Eppendorf is still correct and everyone knows that's not GLP -right? But let's be clear, most people don't give a damn about how many photoelectrons there are -they just want a pretty image. For the few cases where photo-electron numbers are needed, the time taken to take darks and flats are trivial compared to the time taken in precise experiments.
>>>
>>> my 2c
>>>
>>> Mark Cannell
>>>
>>>
>>>
>>>
>>>