Re: Lack of penetration of immunolabelling

Posted by Pertti Panula on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Acceptor-Photobleaching-vs-Sensitized-Emission-FRET-results-tp4828946p4936077.html

Hi,

In particular, with nestin and some other structural proteins, fixation
  may also be an issue. For a number of best available antibodies, PFA
is not optimal, but acetone works well. If needed, please contact prof.
Ismo Virtanen at ismo.virtanen at helsinki.fi for details.

Best regards Pertti



[hidden email] wrote:

>
> Bonjour à tous,
>
> We are trying to (indirect) immunostain with antiboby on  heart
> cryosections (14um).  The proteins that we want to identified are from
> the cytoskeleton like desmin, myosin light chain, myosin heavy chain.
>  We have a problem that all the staining stay on top of the tissue for
> those antibodies.  Counterstaining for actin (phalloidin alexa
> conjugated) is pretty good through the thickness of the tissue.  We
> permealized the tissue with Triton 0.5% , in the blocking solution
> (serum from the host of the 2nd Ab), for 1 hour and we incubate the
> primary overnight at  4C (with Triton 0.2% in the antibody diluent).
>  Other antibodies against cytoskeleton protein (myosin light chain 7)
> are well stained over the thickness of the tissue using the same
> protocol and the same batch of tissues.
>
> Any clue that might help us?
>
> Louissssssss
> Louis Villeneuve
> Research Associate- Confocal Microscopy
> Heart Montreal Institute- Research Center
> 5000 East Belanger
> Montreal (Qc), Canada
> H1T 1C8
>
> 514-376-3330 ext 3511
> 514-376-1355 (Fax)
>
> [hidden email]


--
Pertti Panula
Professor, Vice Dean of Research
Neuroscience Center
Institute of Biomedicine/Anatomy
Faculty of Medicine
POB 63, 00014 University of Helsinki
Finland
Phone: +358 9 19125263
Fax: +358 9 191 25261
Mobile: +358 40 5922 323
pertti.panula at helsinki.fi
http://www.helsinki.fi/neurosci/panula.htm