Re: Photobleach (FRAP) problem in Zeiss 510Meta

Posted by Cselenyak Attila on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Photobleach-FRAP-problem-in-Zeiss-510Meta-tp5038048p5038258.html

Were 50 iterations enough at the previous experiments? Sometimes I
need 100 iterations with 488 laser line at 100% with living cells.
Have you seen any bleaching in the whole image?

Best,

Attila

On Tue, May 11, 2010 at 10:21 PM, G.Liu <[hidden email]> wrote:

> Dear all,
>
> We have encountered a problem in our 510Meta for photobleaching. We simply
> could not get the GFP ROI bleached. What we did is as following, anything
> wrong or inappropriate?
>
>
> 1. Collect the Single unsaturated image as usual (normal settings).
>
> 2. Go to Bleach control for bleach settings. Select xx scans before the
> bleach. For GFP, tried 1-50 iteration, defined the Bleach region. All the
> argon laser lines (458, 488, 514) or only 488 to 100% .
>
> 3. Go to Time Lapse--using Manual start, manual stop (input the number of
> scan). Time interval 0 or 1 sec.
>
> 4. Hit StartB.
>
> I can get a series of images but no bleaching was found. The other odd thing
> is that "Bleach" button seems not not working--if hit, it only lasts a few
> seconds then stop. I've tried using diode405 line, it worked once, but I
> can't repeat it anymore.
>
> Any suggestions or advice are greatly appreciated!!
>
> Guosheng
>
>
>
> --
> View this message in context: http://confocal-microscopy-list.588098.n2.nabble.com/Photobleach-FRAP-problem-in-Zeiss-510Meta-tp5038048p5038048.html
> Sent from the Confocal Microscopy List mailing list archive at Nabble.com.
>