Posted by
John Runions on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Photobleach-FRAP-problem-in-Zeiss-510Meta-tp5038048p5051549.html
!!!! Don't use Krazy Glue or any solvent based glue on leaves. Even at
some distance from the cells, this will kill them!!!
I would suggest removing the actin cytoskeleton using, e.g. latrunculinB,
if nuclear movement is an issue. This, of course depends on what your
experiment is. If you are trying to observe FRAP of a protein or
mechanism that transports on microfilaments, then depolymerising the actin
won't work.
John.
Dear All,
Thanks you very much for all of your input (comments, suggestions,
protocols
etc ) in regard to the photobleaching in our Zeiss510. The good news is
that
we got some progress on this but bad news is that a new question needs to
be
addressed: the movement of the bleaching subject/organelle.
We are using transiently GFP-expressed tobacco leaves and want to bleach
whole or partial nucleus. However the nucleus seems to move at lease about
1-2um in a 5min period. we tried to use larger pinhole and larger ROI but
the plane of focus is still an issue (so only saw partial bleach). Zeiss
specialist suggest to stick the leaves on the slide with krazy glue and
surrounded by something to avoid the leave floating/vibrating (we are
using
inverted microscope). We tried but the movement was still noticeable. Have
some of you meet a similar problem and how did it solved?
Meantime I am want to install a macro that could track the cell movement
(developed by Ellenberg). Any experience on it?
Thank you in advance.
--
View this message in context:
http://confocal-microscopy-list.588098.n2.nabble.com/Photobleach-FRAP-problem-in-Zeiss-510Meta-tp5038048p5051254.htmlSent from the Confocal Microscopy List mailing list archive at Nabble.com.
*********************************
C. John Runions, Ph.D.
School of Life Sciences
Oxford Brookes University
Oxford, UK
OX3 0BP
email:
[hidden email]
phone: +44 (0) 1865 483 964
RunionsÂ’ lab web site
(
http://www.brookes.ac.uk/lifesci/runions/HTMLpages/index.html!)
Visit The Illuminated Plant Cell (
http://www.illuminatedcell.com/ER.html)
Oxford Brookes Master's in Bioimaging with Molecular Technology
(
http://www.brookes.ac.uk/studying/courses/postgraduate/2007/bmt)