Re: Optical slice thickness and number for PSF and deconvolution

Posted by Hans on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Optical-slice-thickness-and-number-for-PSF-and-deconvolution-tp5415740p5427736.html

Dear Sudipta and All,

(Message from a commercial vendor)

On 08/14/2010 06:36 PM, Sudipta Maiti wrote:
> Wait a minute. Since Mark and Guy are involved, there is something useful to
> be learned here, and I was reading this with intent. But the original
> question had two points that I feel were not adequately addressed by either.
Good point!
> First, the number of Z-slices required for imaging the bead: more than the

To record a bead or multi-bead image to derive a PSF from it, it should
be recorded at the Nyquist rate or better. The number of Z-planes should
be sufficient to fully contain the PSF. In a typical confocal case 30+
planes should do it. There are a number of caveats though, see for example:

http://www.svi.nl/RecordingBeads
http://www.svi.nl/PsfFromBeads&highlight=distiller

In short, recording the bead is quite independent of recording the
biological object, provided that optical conditions are the same. If
bleaching or time constraints dictate few planes of the object, then
from the standpoint of deconvolution a slight undersampling in Z is
acceptable. But not too much!

> original can help, as the deconvolution algorithm will probably use a smooth
> function to model the PSF obtained from the subresolution bead image, and use
> that for deconvolution.
> Second, was there something about using a larger pinhole during the actual
> image acquisition? The pinhole size should match for the actual imaging and
> the bead - otherwise you don't get the same PSF. I guess it is possible to
> calculte the PSF for other pinhole sizes, but may not be the best thing to do.
Indeed, the pinhole sizes used in recording the PSF and the data should
match. The sampling can be adapted and the tails of the PSF can be
extrapolated though it is better if this is not necessary.

Regarding the tradeoff between pinhole size, microscope type, signal and
deconvolution, the outcome of that is very much object dependent. As a
rule of thumb, with small sparse objects you're likely to be better off
with a WF system + deconvolution, with thick dense objects with a
confocal system. Increasing the pinhole beyond the Airy disk size gets
you less Z-resolution and more signal in the absolute sense, but as Guy
pointed out more (blurry) signal from adjacent areas carries also more
noise. Deconvolution can repair the resolution loss, again much
depending on the sparseness of the object. For example, if there are
strongly labelled horizontal membranes in your image next to other
interesting but faint features, opening the pinhole is probably not a
good idea.

Regarding the number of planes in the data to be recorded for
deconvolution: if possible record enough planes to contain the entire
object. If that is not feasible, then the more the better.
In the low noise WF sparse object case even single plane deconvolution
is possible (this is not 2D deconvolution!), in the confocal case three
or more Nyquist sampled planes are needed. Of course reliability is not
as good that of a full-object deconvolution, and there are cases where
3-plane confocal 'deconvolution' will fail. But as long as the object is
fairly sparse and there are no large bright objects right outside the
recorded volume, my colleague Vincent's point was that chances are good.
The catch is this: because the volume in the data contributing to a
deconvolved pixel is less in the 3-plane case compared to a full-object
deconvolution, the signal to noise (SNR) requirements are a bit higher
than in the full object case. If the choice is between 3 higher SNR
planes or more lower SNR planes, both cases involving the same amount of
detected photons, best go for more planes.

We've done simulations on this topic in the past, if anyone is
interested we could construct a wiki page.

-- Hans

> Sudipta
>
>      On Sat, 14 Aug 2010 23:02:18 +1000, Guy Cox wrote
>> It's a bit challenging to disagree with Mark but .... we are interested
>> in signal over noise.  Opening the pinhole beyond the diameter of the
>> Airy disk will let in a little more signal (from the outer rings)
>>   and a LOT more noise.  In almost every case it will make things worse.
>> Photons are precious - if they come from where we want - other
>> photons are something we need to exclude at all costs.
>>
>> As to the question about 3 sections - Mark is quite right, of course,
>>   if we are dealing with a thick sample, but if I've followed this thread
>> correctly we are dealing with thin cells where the information is
>> largely in one plane.  If this is so we should be able to do pretty good
>> deconvolution with 3 sections.
>>
>>                                                        Guy
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor&  Francis
>>       http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon)
>> Australian Centre for Microscopy&  Microanalysis,
>> Madsen Building F09, University of Sydney, NSW 2006
>>
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>               Mobile 0413 281 861
>> ______________________________________________
>>        http://www.guycox.net
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]]
>> On Behalf Of Mark Cannell
>> Sent: Friday, 13 August 2010 8:52 AM
>> To: [hidden email]
>> Subject: Re: Optical slice thickness and number for PSF and
>> deconvolution
>>
>> Hi All
>>
>> I'm sorry but this advice is wrong. The pinhole is a control that
>> _should_ be used when decreased (mainly z) resolution is acceptable.
>> The
>>
>> lasers can then be turned down and, if desired, decon. can be used
>> to help clean up the image. The problem is that many users want a
>> "pretty picture" but pretty pictures may not be needed for
>> quantification of
>> (say) number of mitochondria.  As we say on the Vancouver course, "Every
>>
>> photon is precious" and you may also increase signal by accepting a
>> wider spectral band or using an LP filter.  The key to good experimental
>>
>> work is to understand what measurement you want and then to pick
>> conditions that allow you to get sufficient data with sufficient
>> (not too many) time points to answer your question. Do you need a
>> full 3D image or will a couple of slices suffice? Use a high NA
>> lens. As others have said, consider using widefield with a high QE
>> CCD if you really don't need the maximum possible resolution in 3D...
>>
>> My 2c
>>
>> Mark Cannell
>>
>> Vincent wrote:
>>> *commercial interest*
>>>
>>>
>>> Dear Jan,
>>>
>>> The amount of the signal in images is mostly judged just after image
>>> acquisition. Based on this it is often decided to use a wider pinhole.
>>> As you probably know, when deconvolution is properly performed you
>> will gain not
>>> only an increase in resolution but also in signal. Therefore, we
>> advise to close
>>> the pinhole and use deconvolution for increasing the signal (to noise)
>> before
>>> determining the quality of the image.
>>>
>>> As with imaging the object of interest it is important to follow the
>> Nyquist
>>> criteria for imaging the bead images.
>>> We have a Nyquist calculator on our website
>> (www.svi.nl/NyquistCalculator) to
>>> determine these rates. You can also create a picture here of your
>> theoretical
>>> PSF to get an idea of its dimensions.
>>>
>>> In general it is best to really match the Nyquist criterion in xyz.
>> Else you can
>>> go for 2x more. This however may introduce other problems like e.g.,
>> bleaching.
>>> If the bead images are differently sampled it requires interpolation
>> for
>>> matching that, making the process of deconvolution more
>> computationally
>>> demanding. Thus Nyquist is okay. Another important thing to keep in
>> mind is that
>>> you need to image enough planes to cover your PSF.
>>>
>>> I hope this answers your questions.
>>> Best regards,
>>> Vincent
>>>
>>> ***********************************************************
>>> Vincent Schoonderwoert, PhD
>>> Scientific Volume Imaging bv
>>> Hilversum, The Netherlands
>>> [hidden email]
>>> [hidden email]
>>> Tel: + 31 35 646 8216
>>> ***********************************************************
>>>
>>>
>>>
>>>
>>>
>>> Jan Trnka wrote:
>>>> Dear list,
>>>>
>>>> this is probably a trivial question but so far I haven't found a
>> good answer.
>>>> When taking 3D images of subresolution beads in a confocal
>> microscope (for PSF
>>>> construction) does the number and thickness of slices in the z-stack
>> need to
>>>> be exactly the same as that of a sample to be deconvolved? I
>> understand the
>>>> x-y dimensions need to be the same but how does it work for z? Would
>> a higher
>>>> number of thinner slices (finer z resolution) of the bead improve
>> the
>>>> construction of the PSF? My actual samples are imaged with a rather
>> wide
>>>> pinhole setting to limit the exposure of the sample (live cells) and
>> thus
>>>> provide quite thick optical sections.
>>>>
>>>> Thanks,
>>>>
>>>> Jan
>>>>
>>>> Jan Trnka, MD, PhD
>>>> Department of Biochemistry
>>>> 3rd Medical Faculty
>>>> Ruska 87
>>>> 100 00 Praha 10
>>>> Czech Republic
>>>> [hidden email]<mailto:[hidden email]>
>>>> Tel.: +420 26710 2410
>>>>
>>>>
>>>>
>>>
>>>
>>>
>>>
>>
>> No virus found in this incoming message.
>> Checked by AVG - www.avg.com
>> Version: 9.0.851 / Virus Database: 271.1.1/3032 - Release Date: 08/12/10
>> 04:34:00
>
>
> Dr. Sudipta Maiti
> Associate Professor
> Dept. of Chemical Sciences
> Tata Institute of Fundamental Research
> Homi Bhabha Raod, Colaba, Mumbai 400005
> Ph. 91-22-2278-2716 / 2539
> Fax: 91-22-2280-4610
> alternate e-mail: [hidden email]
> url: biophotonics.wetpaint.com
>


--
-------------------------------------------------------------------
dr. Hans T.M. van der Voort                           ([hidden email])
Scientific Volume Imaging b.v.,             URL: http://www.svi.nl/