Hi Tom
Thanks for your suggestion. We use the Zeiss oil for water lens routinely
and it works very well as long as one doesnt try to image a whole multiwell
plate. Oil (normal or for water lenses) is too viscous for large areas. It
collects at places and goes missing at others. We have tried this a lot
without success. The only option we have found to combine higher throughput
and high resolution is to use a water objective. It works really nicely with
glass bottom plates but now we also acquire time lapses, so I guess the only
way to go is to have a water refill system. Hence my query... Does anyone
have any experience with this type of protocols?
Med vänlig hälsning / Best regards
Sylvie
@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader
Live Cell Imaging Unit
Dept of Biosciences and Nutrition
Karolinska Institutet
Sweden
office: +46 (0)8 608 9240
LCI room: +46 (0)8 608 9248
mobile: +46 (0) 73 733 5008
> -----Original Message-----
> From: Donnelly, Tom [mailto:
[hidden email]]
> Sent: 24 September 2010 19:11
> To: '
[hidden email]'
> Subject: RE: water refill device for WI objective
>
> Sylvie;
>
> I have often thought the oil that Zeiss offers that has the same
refractive index of
> water might be a good solution here. Once the plate is coated ( probably
do that
> before starting scan) it would seem that little additional oil would be
needed.
>
> I have no experience with this and have no idea if it is compatible with
Nikon
> objectives, but just a thought.
>
> Tom
>
> Tom Donnelly
> Applied Precision, Inc.
> 1040 12th Ave. N.W. Issaquah, WA 98027-8929
> (425)829-9414 (425)557-1055 fax
>
[hidden email]
>
http://www.appliedprecision.com>
>
>
> -----Original Message-----
> From: Sylvie Le Guyader [mailto:
[hidden email]]
> Sent: Friday, September 24, 2010 6:04 AM
> To:
[hidden email]
> Subject: water refill device for WI objective
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Dear all
>
>
>
> We have a 60x water objective NA1.2 by Nikon that we would like to use to
screen
> 96 and 384 well plates. Does anyone know of a water refill system that
would fit on
> this objective? Or has anyone tried the Olympus or the Leica water refill
device? Do
> they work well? Would any fit our objective?
>
>
>
> Thanks for your help!
>
>
>
> Med vänlig hälsning / Best regards
>
>
>
> Sylvie
>
>
>
> @@@@@@@@@@@@@@@@@@@@@@@@
>
> Sylvie Le Guyader
>
> Live Cell Imaging Unit
>
> Dept of Biosciences and Nutrition
>
> Karolinska Institutet
>
> Sweden
>
> office: +46 (0)8 608 9240
>
> mobile: +46 (0) 73 733 5008
>
>
>
>
> This email message, together with any attachments, is for the sole use of
the
> intended recipient(s) and is the confidential information of Applied
Precision Inc. If
> you are not the intended recipient, your review, use, disclosure, copying
or
> dissemination of this email message or its attachments, or the information
> contained therein, is strictly prohibited. If you are not the intended
recipient or if you
> think this email was sent to you in error, please notify the sender by
reply email and
> delete this message and its attachments, as well as all copies, from your
system.