Re: Pixel size

Posted by Andreas Bruckbauer on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Acceptor-Photobleaching-vs-Sensitized-Emission-FRET-results-tp4828946p5632770.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

The problem is that the magnification of microscope objectives can vary a bit from one to another (same make and type). According to one manufacturer this can be +/- 5% for objectives with correction collar and +/- 2% without. Best to measure it but the theoretical pixel size is a good guess if you are sure about the additional magnification optics.
 
best wishes

Andreas


 

 

-----Original Message-----
From: Christophe Leterrier <[hidden email]>
To: [hidden email]
Sent: Wed, 13 Oct 2010 20:03
Subject: Re: Pixel size


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



If there is no other magnifying/demagnifying optics in your

microscope, the theoretical pixel size is 4.65/40=116.25 nm



Christophe



On Wed, Oct 13, 2010 at 20:43,  <[hidden email]> wrote:

> *****

> To join, leave or search the confocal microscopy listserv, go to:

> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

> *****

>

> Bonjour à tous,

>

> I have a tiff image, 1392 x 1040 - 4.65 um square pixels acquired with a

> camera interline Sony , 1.4 megapixel, color(7.6mm x 6.2mm array).  I use

> a 40 x objective mounted on  a table microscope Leica DME.

>

> Can I find the pixel size in hte image?

>

> Thanks ,

>

> Louis

>