And for the more adventurous types out there, you can measure your penetration depth using fluorescence correlation spectroscopy (FCS) if you have the right hardware:
Harlepp, S., et al., Subnanometric measurements of evanescent wave penetration depth using total internal reflection microscopy combined with fluorescent correlation spectroscopy. Applied Physics Letters, 2004. 85(17): p. 3917-3919.
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> There is also a simple method in Fiolka et al. Miocroscoc. res Tech 2007 :
>
http://www.neuro.nano-optics.ethz.ch/publications/fiolka.pdf>
> <
http://www.neuro.nano-optics.ethz.ch/publications/fiolka.pdf>involving an
> inclined coverslip with attached fluorescent beads. I didn't try it myself
> so I can't tell if it works, but seems simpler than AFM or microtubules
> method.
>
> Christophe
>
> On Wed, Nov 24, 2010 at 16:32, John Oreopoulos
> <
[hidden email]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> Sebastian, I remember reading a publication from about ten years ago that
>> talked about mounting a pencil in a piezoelectric micro-manipulator and
>> sticking a fluorescent bead on the end of the pencil tip. This was very
>> similar to the AFM method. Is it the same thing?
>>
>> Graham, there is a fairly detailed discussion on this topic from a few
>> years ago on the archive that talks about a few other ways to measure the
>> actual TIRF penetration depth (as opposed to calculating it based on an
>> assumed refractive index and crudely measured incident angle):
>>
>>
>>
http://lists.umn.edu/cgi-bin/wa?A3=ind0702&L=CONFOCALMICROSCOPY&E=quoted-printable&P=827770&B=--Apple-Mail-181-588547976&T=text%2Fhtml;%20charset=ISO-8859-1>>
>> I've tried the AFM method as well - it works, but my main complaint with
>> this and some of the other protocols is that they require complicated and
>> expensive equipment, and can be difficult to get right. The method involving
>> fluorescent microtubles by Jorg Enderlein's group is a fairly new one that
>> is elegantly simple, but again requires you to have access to some fairly
>> special reagents that might not be found in every lab. A few months ago I
>> came across yet another older method that had evaded my previous searches on
>> the topic. This one is similar to the others that involve imaging
>> fluorescent microbeads, but I like this because all it requires is a
>> microscope with a motorized drive on the z-axis:
>>
>> Steyer, J.A. and W. Almers, Tracking single secretory granules in live
>> chromaffin cells by evanescent-field fluorescence microscopy. Biophysical
>> Journal, 1999. 76(4): p. 2262-2271.
>>
>> John Oreopoulos
>> Research Assistant
>> Spectral Applied Research
>> 9078 Leslie Street, Unit 11
>> Richmond Hill
>> Ontario, Canada
>>
>>
>> On 2010-11-24, at 6:54 AM, Sebastian Rhode wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>> *****
>>>
>>> Hi Graham,
>>>
>>> one method which definitly works (I tried it out by myself) is the use of
>> an
>>> combined TIRF-AFM setup. You just have to couple fluorescent beads to the
>>> tip of your AFM an record pictures while approaching/or moving away the
>>> coverslip surface. Unfortunately an AFM is really expensive.
>>>
>>> So I found some other methods, which might work as well --> see
>>> TIRF_Introduction.pdf, which I send to you directly (LIST server does not
>>> accepted this pdf-file).
>>>
>>> One methodes uses an objective piezo-drive and a pencil and the second
>> one
>>> stained beads or a stained solution with intransparent beads.
>>>
>>> I case of questions, feel free to contact me directly.
>>>
>>> Cheers,
>>> Sebastian
>>>
>>>
>>> Dr. Sebastian Rhode
>>> Project Manager
>>> Research & Development
>>>
>>> TILL Photonics GmbH
>>> Lochhamer Schlag 21
>>> D- 82166 Gräfelfing, Germany
>>> Phone +49 (0)89 895 662-120
>>> Fax +49 (0)89 895 662-101
>>> www.till-photonics.com
>>