Re: TIRF depth calibration

Posted by Lauran Oomen on
URL: http://confocal-microscopy-list.275.s1.nabble.com/TIRF-depth-calibration-tp5769934p5773696.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Another publication related to this topic is this one:
http://spiedl.aip.org/getabs/servlet/GetabsServlet?prog=normal&id=JBOPFO000011000001014006000001&idtype=cvips&gifs=yes&ref=no
Mattheyses, A.L. and Axelrod, D.
Direct measurement of the evanescent field profile produced by objective-based total internal reflection fluorescence
J. Biomed. Opt., Vol. 11, 014006 (2006)

Did not try it (yet), however.

Best wishes

Lauran Oomen
--------------------------------------------------------------
Lauran Oomen
Manager Digital Microscopy Facility (C.2.023)
NKI-AVL
Plesmanlaan 121
PO Box 90203
1006 BE Amsterdam
The Netherlands
 
phone +31 205126080
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of John Oreopoulos
Sent: woensdag 24 november 2010 17:21
To: [hidden email]
Subject: Re: TIRF depth calibration

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

And for the more adventurous types out there, you can measure your penetration depth using fluorescence correlation spectroscopy (FCS) if you have the right hardware:

Harlepp, S., et al., Subnanometric measurements of evanescent wave penetration depth using total internal reflection microscopy combined with fluorescent correlation spectroscopy. Applied Physics Letters, 2004. 85(17): p. 3917-3919.

John Oreopoulos
Research Assistant
Spectral Applied Research
9078 Leslie Street, Unit 11
Richmond Hill
Ontario, Canada


On 2010-11-24, at 10:37 AM, Christophe Leterrier wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> There is also a simple method in Fiolka et al. Miocroscoc. res Tech 2007 :
> http://www.neuro.nano-optics.ethz.ch/publications/fiolka.pdf
>
> <http://www.neuro.nano-optics.ethz.ch/publications/fiolka.pdf>involving an
> inclined coverslip with attached fluorescent beads. I didn't try it myself
> so I can't tell if it works, but seems simpler than AFM or microtubules
> method.
>
> Christophe
>
> On Wed, Nov 24, 2010 at 16:32, John Oreopoulos
> <[hidden email]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Sebastian, I remember reading a publication from about ten years ago that
>> talked about mounting a pencil in a piezoelectric micro-manipulator and
>> sticking a fluorescent bead on the end of the pencil tip. This was very
>> similar to the AFM method. Is it the same thing?
>>
>> Graham, there is a fairly detailed discussion on this topic from a few
>> years ago on the archive that talks about a few other ways to measure the
>> actual TIRF penetration depth (as opposed to calculating it based on an
>> assumed refractive index and crudely measured incident angle):
>>
>>
>> http://lists.umn.edu/cgi-bin/wa?A3=ind0702&L=CONFOCALMICROSCOPY&E=quoted-printable&P=827770&B=--Apple-Mail-181-588547976&T=text%2Fhtml;%20charset=ISO-8859-1
>>
>> I've tried the AFM method as well - it works, but my main complaint with
>> this and some of the other protocols is that they require complicated and
>> expensive equipment, and can be difficult to get right. The method involving
>> fluorescent microtubles by Jorg Enderlein's group is a fairly new one that
>> is elegantly simple, but again requires you to have access to some fairly
>> special reagents that might not be found in every lab. A few months ago I
>> came across yet another older method that had evaded my previous searches on
>> the topic. This one is similar to the others that involve imaging
>> fluorescent microbeads, but I like this because all it requires is a
>> microscope with a motorized drive on the z-axis:
>>
>> Steyer, J.A. and W. Almers, Tracking single secretory granules in live
>> chromaffin cells by evanescent-field fluorescence microscopy. Biophysical
>> Journal, 1999. 76(4): p. 2262-2271.
>>
>> John Oreopoulos
>> Research Assistant
>> Spectral Applied Research
>> 9078 Leslie Street, Unit 11
>> Richmond Hill
>> Ontario, Canada
>>
>>
>> On 2010-11-24, at 6:54 AM, Sebastian Rhode wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hi Graham,
>>>
>>> one method which definitly works (I tried it out by myself) is the use of
>> an
>>> combined TIRF-AFM setup. You just have to couple fluorescent beads to the
>>> tip of your AFM an record pictures while approaching/or moving away the
>>> coverslip surface. Unfortunately an AFM is really expensive.
>>>
>>> So I found some other methods, which might work as well --> see
>>> TIRF_Introduction.pdf, which I send to you directly (LIST server does not
>>> accepted this pdf-file).
>>>
>>> One methodes uses an objective piezo-drive and a pencil and the second
>> one
>>> stained beads or a stained solution with intransparent beads.
>>>
>>> I case of questions, feel free to contact me directly.
>>>
>>> Cheers,
>>> Sebastian
>>>
>>>
>>> Dr. Sebastian Rhode
>>> Project Manager
>>> Research & Development
>>>
>>> TILL Photonics GmbH
>>> Lochhamer Schlag 21
>>> D- 82166 Gräfelfing, Germany
>>> Phone   +49 (0)89 895 662-120
>>> Fax     +49 (0)89 895 662-101
>>> www.till-photonics.com
>>