Re: DsRed vs 2P

Posted by Craig Brideau on
URL: http://confocal-microscopy-list.275.s1.nabble.com/DsRed-vs-2P-tp5830859p5839979.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

That's a very good point!  I found from my own compressor that the longer
you tune the less sensitive the pulse is to dispersion.  This may account
for some of the brightening seen as the laser was tuned more to the
infrared!  Thanks for bringing that up and jogging my memory.
There are so many factors to consider when characterizing these things!

Craig


On Wed, Dec 15, 2010 at 5:30 AM, Sudipta Maiti
<[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Remember that the pulsewidth will also be changing as you tune, so if you
> have
> a pulse compressor, you may want to tweak that too to keep it constant.
> Though
> the dependence for 2P is linear, pulsewidths can change quite drastically
> towards the ends of the tuning range of a Ti:sapph.
> Sudipta
> On Wed, 15 Dec 2010 06:38:48 -0500, Andreas Bruckbauer wrote
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> >  The pick up from the laser is a good idea and i think microscope
> > manufacturers should implement this and get rid of the percentage
> > numbers for laser powers which are widely used. However this will
> > not account for the differences in transmission of the different
> > optical elements in the microscope. So the Zeiss macro is a good
> > idea if you are interested in the fluorophore characteristics. The
> > only thing to keep in mind is that, if your sample allows you to use
> > more power and you are operating at 100%,  you still might gain from
> > using the lower wavelength. The fluorescence goes with the square of
> > the laser power, so 2.5x more power translates to 6x more
> > fluorescence, if the fluorescent protein is 3x more efficient at the
> > longer wavelength (at same power level), you would still get 2x more
> > fluorescence using the lower wavelength at higher power. But these
> > are just theoretical values.
> >
> > best wishes
> >
> > Andreas
> >
> > -----Original Message-----
> > From: Paul Herzmark <[hidden email]>
> > To: [hidden email]
> > Sent: Tue, 14 Dec 2010 22:20
> > Subject: Re: DsRed vs 2P
> >
> > *****
> >
> > To join, leave or search the confocal microscopy listserv, go to:
> >
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >
> > *****
> >
> > Using my Tai-Saphire laser, I measured the optimal 2P excitation
> wavelength
> >
> > for TD Tomato (a different red fluorescent protein).
> >
> > Since the wattage coming out of the laser varies with wavelength you
> > need to
> >
> > correct for that. I have a pickoff after my laser power attenuator
> > that goes
> >
> > to a power meter so I can keep the power delivered to the cells constant.
> >
> > Scan through the wavelengths changing the attenuator to keep the power
> >
> > constant and look for the brightest emission. I found 1040 nm (the
> > end of
> >
> > what my laser will deliver) to be about 5X more efficient per watt
> > than 920
> >
> > nm.
> >
> >  It was still going up when I got to 1040 so I am going to try an OPO
> >
> > (Optical Parametric Oscillator) to look at even longer wavelengths.
> >
> > Keeping the delivered power constant and scanning the spectrum is
> > also a
> >
> > good way to find the optimum compromise power for simultaneously exciting
> >
> > several fluorophores.
> >
> > Paul
> >
> > Paul Herzmark
> >
> > Specialist
> >
> > [hidden email]
> >
> > Department of Molecular and Cell Biology
> >
> > 479 Life Science Addition
> >
> > University of California, Berkeley
> >
> > Berkeley, CA  94720-3200
> >
> > (510) 643-9603
> >
> > (510) 643-9500 fax
> >
> > 2010/12/13 Staffan Nyström <[hidden email]>
> >
> > > *****
> >
> > > To join, leave or search the confocal microscopy listserv, go to:
> >
> > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >
> > > *****
> >
> > >
> >
> > > Dear list members,
> >
> > >
> >
> > > Has anyone any experience with visualizing DsRed using Ti:Saph (2
> photon)?
> >
> > > I have problems getting any decent fluorescence signal through between
> 700
> >
> > > and 1000nm excitation and there is significant bleedthrough
> >
> > > with EGFP. In the end I gave up since the "regular" 1 photon excitation
> way
> >
> > > gave a much better S/N ratio with minimal bleed through.
> >
> > >
> >
> > > Any tips? We have an upright Zeiss LSM system and mostly use 40X 1.0 NA
> >
> > > water dipping objectives.
> >
> > >
> >
> > > Best
> >
> > >
> >
> > > Staffan Nystrom
> >
> > >
> >
> > > PhD student
> >
> > >
> >
> > > MBB / Oncology & Pathology
> >
> > >
> >
> > > Karolinska Institutet
> >
> > > Stockholm, Sweden
> >
> > >
> >
> > =
>
>
> Dr. Sudipta Maiti
> Associate Professor
> Dept. of Chemical Sciences
> Tata Institute of Fundamental Research
> Homi Bhabha Raod, Colaba, Mumbai 400005
> Ph. 91-22-2278-2716 / 2539
> Fax: 91-22-2280-4610
> alternate e-mail: [hidden email]
> url: biophotonics.wetpaint.com
>