Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Thanks all for reply.

Given that we provide 1/2 hour for laser to settle is it reasonable to expect that one can tune the laser thrice a day for different users and still it works fine? Autocorrelator is too costly for time being but we do have a spectrum analyzer. Upto now, the optical setups have been aligned with aid of only the spectrum analyzer.

Is there a good application note/ reading available that explains the light-path of MIRA 900 (both picosecond and femtosecond configurations) in context of Kerr Lens Modelocking? - I don't know about either. Also some 'protocol' of tuning will be great. We have a manual from Coherent, but I am finding it hard to follow.

Regards
Shalin

On 8/30/07, Guy Cox <[hidden email]> wrote:
Search the CONFOCAL archive at <a href="http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Shalin wrote;
 
 What diagnostics do you use for tuning of two photon laser or OPO? Is laser spectrum analyzer enough?
The laser spectrum analyser should get you back exactly to your
wavelength, and give a pretty good idea of the pulse length.
Lastek in Australia make a software system which works with a
spectrum analyser to give you pulse length and it seems pretty
good provided it is carefully calibrated.  (It's probably sold under
some other name in the States).
 
But if you have users who want to know the pulse length precisely
it's really worthwhile buying an autocorrelator. 
 

We have a shared two photon (MIRA 900F Femtosecond configuration) and OPO from Coherent. One of the main user is very afraid of tuning it - so he always has a service engineer tune it to 830nm and longest possible pulse (since he does spectroscopy). There are two more of us who will benefit by tuning it for two-photon and SHG (different wavelengths and shortest possible pulse).
 
830 is quite OK for SHG (brings the harmonic to 415 which is within
the range of conventional interference filters (they start to absorb a lot
at anythong shorter).  But it's a bit limiting for 2P.

How is the 'reproducibility' of laser beam profile when working with these lasers? One of the main concern that other user has is that - if you cannot come back to exactly 830 nm after tuning to say 740 nm, the custom optical setup will go out of alignment.
In the early days one make of Ti-S laser was notorious for beam
wander as you tuned it but even so it should go back when you
returned it to the original wavelength.  Nowadays they seem really
stable, and I don't think your Mira should be a problem there.  The
only thing I find with a Mira is that it's  best to let it warm up for
half an hour before doing any tweaking of the alignments (and often
that way you find you didn't really need to do any tweaking anyway).
 
                                                                                            Guy
 



--