Hi Stephen,
First, consider some wetware changes:
CFP (presumably ECFP?) -> mCerulean ... Piston lab
YFP (presumably EYFP?) -> mVenus ... Miyawaki lab
mRFP (presumably mRFP1?) -> mKate ... just published by Scherbo et al
in Nature Methods (PubMed ID 17721542)
The mKate article compares all the far red fluorophores.
camera tips:
1. have you optimized (maximized) the gain setting on the ORCA? Your
software default may be low gain. If you change the gain, you may also
need to tweak the offset to keep the 'floor' above zero (I like having it
at about 100 or 200 intensity levels). There are 10 gain settings on the
ORCA-ER Firewire cameras. These are analog gains, before the digitizer,
so output is continuous.
Given your exposure times (3 seconds!) I suspect you are
using low gain.
2. The ORCA-ER Firewire has an "IR mode", that is useful for
far-red/near-infrared signals (on for mRFP1 or mKate or Cy5, off for
visible fluorophores). It increases the QE in the red by grabbing
photoelectons from deeper in the CCD (blue is absorbed first, then green,
then red, then IR), at the cost of a slightly higher background (more
volume of the CCD, results in also grabbing more thermal electrons, so
keep it off for blue, green, yellow).
Zeiss AxioVision and MetaMorph (and I'm sure Compix) users can enable
these options in macros/journals, as well as in the acquire dialogs.
My thanks to Butch Moomaw of Hamamatsu for explaining these
features.
emission filters: do you have spectrally optimal emission filters? Are
the filters in excellent shape (no fingerprints, no pinholes)?
illumination: Assuming you are using laser(s), are you getting optimum
output from them to the objective lens? I'm guessing you are using an
Argion ion laser for the CFP and YFP - is its power output (knob or
slider for Amperage, not AOTF) adjusted for good power output? Too low
Amperage results in instability, too high results in shortened tube
lifetime (max power can result in 1/10th tube life - of course if the
tube is under service contract and/or the experiments are worth it,
cranking the Amperage may get you better data).
Photobleaching: if you have higher illumination, you may (should!) have
more photobleaching. Oxyrase
(
www.oxyrase.com)
or other additives have been used to decrease O2
New camera and/or new lenses: Get demos!!!
Lenses: besides (getting demo of) the 1.45 NA TIRF lenses, operated in
TIRF and non-TIRF modes, how well does the Zeiss 63x/1.2 NA water
immersion lens work for your live cell experiments? If you are
micrometers away from the coverglass, RI matching MIGHT be preferable to
high NA.
If you want to go faster, simultaneous acquisition of the three emission
channels is possible with multiple cameras or the dualview/quadview from
Optical Insights/Photometrics (some other companies have side-by side
devices, and see Prabhat et al 2007 PNAS for an interesting configuration
for a TIRF/widefield system).
Vitaly wrote in another response: TIRF may not work as you would excite
molecules which are up to 200 nm away from the glass. Moreover TIRF
people often use a thicker cover glass to reduce optical
artefacts.
Why would a thicker coverglass reduce optical artifacts in TIRF?
Measuring the coverglass to make sure it is correct for the emission side
(i.e. 170 um for a standard 0.17 coverglass corrected lens) should be
what matters. The TIRF excitation is propagating inside the coverglass at
the coverglass-cell culture medium interface - how would it know how
thick the coverglass is to the objective lens side?
best wishes,
At 12:38 PM 8/31/2007, you wrote:
Search the CONFOCAL archive
at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear microscopists,
Here's the situation:
I need to reduce my exposure times.
I am visualizing dynamically moving complexes
(~150nm/s) that are
composed of low abundance signaling molecules, and are close to the
coverslip-water interface.
I need to visualize these structures, in multiple
colors, every 3-5
seconds- preferably faster. Presently, we use a Yokogawa spinning disc
to
detect CFP, YFP, and mRFP variants. We are getting by, but would like to
be
able to do the work at more physiological chimera expression levels.
Typical
exposures currently range from 500 ms (good) to 3000 ms (bad) per
channel.
I am achieving adequate resolution with a 40X NA1.3
oil immersion lens
(Zeiss). I am using the Hamamatsu ORCA-ER CCD. The camera has 6µm
pixels,
which we use unbinned. I do not think I can tolerate any lower
resolution.
I have two options, and a question associated with
each:
(1) Move to a TIRF system. What are your opinions
about how much gain in
sensitiviy this may provide? Has anyone worked with the Zeiss TIRF
module
for the Axiovert 200M?
(2) Change cameras. I have tried a few back thinned
EM-CCDs, but did not
find that they offered much benefit once the pixel size was corrected
for.
Is there a better option, that offers high resolution, high sensitivity,
and
low background noise? Obviously, there will be compromises. What are
your
opinions of intensified CCD cameras?
Best regards,
-Steve Bunnell
****************************************************************************
Stephen C. Bunnell, Ph.D.
Assistant Professor
Tufts University Medical School