Re: Estimating field of view in an LSM

Posted by Barbara Foster on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Estimating-field-of-view-in-an-LSM-tp590288p590289.html

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Craig

This is one of the easiest and most useful of calculations:

The full field of view of an objective is dependent on two factors:  the magnification of the objective and the FIELD NUMBER of the eyepiece.  The field number of the eyepiece is determined by the distance across a baffle which is about about 2 cm up, inside the bottom of the eyepiece. However, you don't need to measure it:  the value, in mm, is engraved on the eyepiece. Older or simpler microscopes typically have field numbers on the order of 18mm.  Newer microscopes and research stands typically range from 22mm to 25 mm; ultra wide eyepieces may go as high as about 35mm.

Please note that I said "full field of view" because the field aperture iris (located at the light port in the base of the microscope can be used to decrease the size of the field of view. This control is valuable if you have highly scattering features:  simply move the object of interest to the center of the field and close the image of the field iris around it. For most samples, you will be surprised at how much clearer and crisper the image becomes... all without any loss of resolution. 

As for the calculation:  Field of view = field number/magnification of the objective*

Note the asterisk.  The if there is any other INTERMEDIATE magnifier, multiply the Mobj x Mintermediate.  The eyepiece is NOT part of this equation. 

On a more empirical note, you can actually measure the REAL diameter of the field of view directly using a stage micrometer.  This approach works fairly well for everything except the very lowest magnifications.  In that instance, you need some reference point in the eyepiece (cross hair or an eyepiece micrometer).  Align the stage micrometer to one side of the field and take a reading to the reference point then align the stage micrometer to the other side of the field and measure to the reference point.  Simply add the two values together.  This approach is especially useful with cameras because the addition of the phototube and the camera chip frequently reduce the field of view seen through the microscope. 

Hope this was helpful.

Best regards
Barbara Foster, President

We've moved!
Microscopy/Microscopy Education
7101 Royal Glen Trail, Suite A
McKinney TX 75070
P: (972)924-5310
Skype: fostermme
W: www.MicroscopyEducation.com


MME is now scheduling customized, on-site courses through December.  Call us today for details.

P. S.
Need a good general reference or light microscopy text for next semester? Call us today to learn more about "Optimizing LIght Microscopy".  Copies still available through MME... even for class-room lots ... and we give quantity discounts. Just call us here in the MME office for details.




At 03:00 PM 9/23/2007, Craig Brideau wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I was wondering how I estimate the field of view for a given lens in a
laser scanning microscope?  I know it will depend on the entrance
angle of the beam into the back aperture of the lens, but I don't know
how to translate this into lateral focal point displacement for a
given working distance/NA/magnification of lens.  Can anyone point me
towards some good references?

Thanks,

Craig