Re: Nikon C1si?

Posted by Zucker.Robert on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Nikon-C1si-tp590408p590413.html

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Criag
If spectral analysis is the endpoint you may want to consider the PARISS
system from Lightform. ( http://www.lightforminc.com/) that can be
placed on a widefield microscope.
We are using the PARISS system to obtain spectral data --it has 1nm
resolution and has a spectrum from 400-900nm. I find it to be a great
asset in the laboratory,
It is far more sensitive and accurate that any confocal system that I
have seen.

Contact Jeremy Lerner for details on his system.
best wishes
Bob

                                                                             
                                                                             
                                                                             
                                                                             
                                                                             
                                                                             
                                                                             
                                LightForm, Inc.,                            
                          601 Route 206, Suite 26-479                        
                             Hillsborough NJ 08844                          
                               Tel: (908)281 9098                            
                        Email: [hidden email]                      
                                                                             
                                                                             
                                                                             



.

Robert M. Zucker, PhD
U.S. Environmental Protection Agency
Office of Research and Development
National Health and Environmental Effects Research Laboratory.
Telephone: 919-541-1585   Fax: 919-541-4017
e-mail: [hidden email]

Mail address: Reproductive Toxicology Division, MD 67
2525 E.NC Highway 54
Research Triangle Park, North Carolina, 27711

Shipping address: 2525 E.NC Highway 54
Durham, NC, 27713



                                                                       
             Craig Brideau                                              
             <craig.brideau@G                                          
             MAIL.COM>                                               To
             Sent by:                 [hidden email]    
             Confocal                                                cc
             Microscopy List                                            
             <CONFOCAL@LISTSE                                   Subject
             RV.BUFFALO.EDU>          Re: Nikon C1si?                  
                                                                       
                                                                       
             10/05/2007 05:48                                          
             PM                                                        
                                                                       
                                                                       
              Please respond                                            
                    to                                                  
                 Confocal                                              
             Microscopy List                                            
             <CONFOCAL@LISTSE                                          
             RV.BUFFALO.EDU>                                            
                                                                       
                                                                       




Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Thanks for your input!  As in everything, there's a trade-off.  Still,
for us the spectral data will be worth it.

Craig

On 10/5/07, Robert Zucker <[hidden email]> wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Craig
> The efficiency of detection is much less in the spectral mode due to
the
> design of the multianode detector. In addition as Kurt Thorn said the
> light is split into a number of different bandpass channels instead of
> being detected by only one channel which also limits the detection in
> each channel.
> Nikon compensates for this decreased light by increasing the pixel
dwell
> time. However photons are photons. With less photons you will get a
> noisier image.
> The spectral detection of any confocal system using a multianode
> detector will not be as good as a PMT designed for good sensitivity
and
> low light detection. It will produce nosier images but if there is
> enough light you will be able to determine a valuable spectrum which
can

> be used for scientific experiments and to manipulate images. Like
> everything with confocal microscopy there is trade-offs and no perfect
> system.
> Best wishes.
> Bob
>
> Robert M. Zucker, PhD
> U.S. Environmental Protection Agency
> Office of Research and Development
> National Health and Environmental Effects Research Laboratory.
> Telephone: 919-541-1585   Fax: 919-541-4017
> e-mail: [hidden email]
>
> Mail address: Reproductive Toxicology Division, MD 67
> 2525 E.NC Highway 54
> Research Triangle Park, North Carolina, 27711
>
> Shipping address: 2525 E.NC Highway 54
> Durham, NC, 27713
>
>
>
>
>              Craig Brideau
>              <craig.brideau@G
>              MAIL.COM>
To
>              Sent by:                 [hidden email]
>              Confocal
cc
>              Microscopy List
>              <CONFOCAL@LISTSE
Subject

>              RV.BUFFALO.EDU>          Re: Nikon C1si?
>
>
>              10/04/2007 02:49
>              PM
>
>
>               Please respond
>                     to
>                  Confocal
>              Microscopy List
>              <CONFOCAL@LISTSE
>              RV.BUFFALO.EDU>
>
>
>
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I was wondering if you saw this decreased throughput in spectral mode
> only, or also in conventional detection mode?  The spectral mode works
> with a multi-anode PMT, which has a lower quantum efficiency than a
> conventional PMT.  From what I have seen of the design, it also has
> the option to use a conventional PMT system with the device, and I am
> wondering if you have also tried that mode of operation?
>
> Thanks,
>
> Craig
>
> On 10/4/07, Robert Zucker <[hidden email]> wrote:
> > One of the major limitations that we have seen with this unit and
also
> > with  the Zeiss meta 510 is the decreased light throughput . This
> > creates images that are noisier than conventional confocal
> microscopes.
> > In our hands it appears you will need to have a bright sample to
make
> > the spectral system work properly. ,
>