Re: Electrical interfacing of external PMT to ADC of FV300

Posted by Guy Cox on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Electrical-interfacing-of-external-PMT-to-ADC-of-FV300-tp590482p590483.html

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
When Nuno Moreno and I did this on a Biorad MRC 1024, we
found we needed a little preamp to convert the negative, current
based signal to a positive voltage-based one.  Dino Sharma of
Biorad kindly gave us the necessary component so we didn't
have to build it.  May be either Dino or Nuno can add some
further details.
 
                                                                             Guy 
 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net


From: Confocal Microscopy List on behalf of Shalin Mehta
Sent: Sat 6/10/2007 10:56 PM
To: [hidden email]
Subject: Electrical interfacing of external PMT to ADC of FV300

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,

We want to add external PMT to FV300 for having non-descanned path. We have old PMT from electron tubes (9798B with QL30 housing). We are a bit puzzled with its behaviour... will be great help if more experienced people could throw some 'light' :-)

We want to match the output of our PMT to ADCs of olympus so that we can interface the PMT using intermediate pre-amp circuit.
So we tested the PMT on the inverted confocal (Olympus FV300) by placing it right on top of the sample and blocking the laser by putting two filters before PMT window. We connected the signal connector (BNC) of the PMT housing to CRO (DC coupling 1 Meg impedance). We simultaneously compared the output of olympus PMT on CRO. The sample were fluorescent beads.

Now we observe this:
1. Output from olympus PMT+ pre-amp circuit was in range of 0-4 V when images appeared under and over saturated on fluoview software. So we concluded that olympus ADCs have a dynamic range of 0-4V.  Output of our PMT is negative voltage (as expected due to electrons flowing from anode to ground).
2. Our PMT does follow increase in fluorescence as sample is scanned at the same speed as olympus PMT (when we look at signals on CRO). But our PMT cannot follow the decrease in fluorescence!!

Our PMT outputs negative pulse and olympus PMT outputs positive pulse.  We see that our PMT 'peaks up' in negative direction exactly in same time as the olympus PMT. But on the opposite path the circuit is slow. To me it seems the problem is with way we are measuring output of PMT. Perhpas the thing is that current sees low time constant circuit when charging but long time constant circuit on the opposite path.

I could send the image and pics if someone thinks will help in making judgement.

could someone please comment on following questions?

1. I think PMT is a current source with internal capacitance and internal resistance connected in parallel. When we connect the CRO (1 Meg) to the housing we are putting a large resistance in parallel. Is that right to assume?

2. If above is the case, the current from sees low time constant path (low internal reistance (from cathode to anode to ground) times capacitance). But on the return path the internal resistance is high (I believe PMT is quite resistant to electron flow from anode to cathode). Since this leaves no return path with small time constant the circuit doesn't come to zero quickly.

3. Perhaps we can design a simple op-amp interving amplifier to provide isolation from CRO and match resistances correctly. If we have equivalent electrical circuit of the PMT we can do it quickly. Anyone has 'typical' internal resistance and capacitance values for QL30 housing? I searched Electron tubes website to no avail.

Any other thoughts on matching external PMT to the ADC of confocal?

As always, thanks for all the inputs,
Shalin



--
My co-ordinates:
Shalin Mehta, Graduate student
Graduate Programme in Bioengineering, NUS, Singapore
Email: shalin {dot} mehta {at} gmail {dot} com
Blog: electricsbm.blogspot.com
Mobile: +65 90694182