http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocalOur Olympus x100's have smaller back apertures and achieve higher effective NA's. If yours are the same, that may explain some of your results.
Mike
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Stanislav Vitha
Sent: Monday, October 29, 2007 11:26 AM
To:
[hidden email]
Subject: Re: Objective for confocal
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocalDear Cameron,
I will be inetrested in the test report on the two Olympus lenses. Were
your tests done in the wide-field or confocal mode?
Our Olympus confocal had a 60x/1.42 PLAPON objective that turned out to be
substantially worse in resolution than we expected from theoretical
calculations. After testing several more of the same type we switched to
the 100x/1.4 Universal Plan Apo (Thanks to our local Olympus office I was
able to pick the best one among several of those objectives). It showed
much better resolution and less chromatic aberration across the visible
spectrum. I can e-mail test results to those who are interested (both
mirror slide and sub-resolution beads were used).
Regarding the signal intensity - I thought that in the point-scanning
confocal mode, there should not be a big difference in excitation
intensity and signal detected between different magnification objectives
of the same NA - the size of the illuminated spot is determined by NA, as
well as the amount of fluorescence signal collected. Since all the
collected fluorescence signal is focused to a single spot in the pinhole
plane, there again should not be much difference between, say, 60x and
100x objective, providing the pinhole size is set to the same Airy size
and all other things are the same. Am I missing something fundamental?
Stan
Dr. Stanislav Vitha
[hidden email]
Microscopy and Imaging Center
Texas A&M University
BSBW 119
College Station, TX 77843-2257
tel: 979-845-1129 (main desk)
tel: 979-845-1607 (direct link)
fax: 979-847-8933
On Sat, 27 Oct 2007 11:08:29 +1000, Nowell, Cameron
<
[hidden email]> wrote:
>Search the CONFOCAL archive at
>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>
>Hi Ann,
> To follow up from Julio's comments. The higher NA will give
you improved resolution. A 1.42 NA object theoreticatly can resolve down
to 210nm at 488nm exitation. That holds true for what ever the
magnification the objective has.
>
>We have compared a 63x 1.42 NA to a 100x 1.42 NA objective. Both were
Olympus UIS2 PlanApo objectives. Both could resolve the same structures.
But the big difference was the amount of light that was needed to capture
an image with the 100x objective. On average 7x more light input was
required to get image with the 100x lens that was equivalent to the 63x
lens.
>
>So my advice woudl be to get a 63x 1.42 NA objective, or better yet a 40x
1.4NA if you can find one (they are being made by Nikon i think but not
sure about others).
>
>Of course this is only important if you want to b e able to resolve very
small structures. If you do not want to be resolving down to 210nm then
your 40x oil and 63x water imersion objective will be just fine.
>
>I can send you (and anyone else on the list) a copy of the reoprt with
the comparison of the 63x vs the 100x lens if you want.
>
>
>Cheers
>
>Cam
>
>
>Cameron Nowell B.Sc (Hons)
>Microscopy Imaging and Research Core Facility Peter MacCallum Cancer
Centre
>7 St Andrews Place
>East Melbourne, Victoria 3002
>
>Phone: +61396561243
>Mobile: +61422882700
>Fax: +61396561411
>
>________________________________
>
>From: Confocal Microscopy List on behalf of Yang, Ann-Fook
>Sent: Sat 27/10/2007 6:09 AM
>To:
[hidden email]
>Subject: Objective for confocal
>
>
>
>Search the CONFOCAL archive at
>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>
>Hi All,
>
>I am new in confocal microscopy.
>We have a 40x/1.3 oil and a 63X/1.2 water objectives. Is there any
advantage to add a 100X/1.46 Oil objective?
>
>Ann Fook Yang,
>Agriculture and Agri-Food Canada/Agriculture et Agroalimentaire Canada
>EM Unit/ Unite EM
>Edifice K.W. Neatby Building,
>960 Carling Av /960 Boulevard Carling,
>Ottawa,Ontario
>K1A 0C6
>
>
[hidden email]
>Telephone/Téléphone: 613-759-1638
>Facsimile/Télécopieur: 613-759-1701
>
>
>
>
>This email (including any attachments) may contain
>confidential and/or legally privileged information and is
>intended only to be read or used by the addressee. If you
>are not the intended addressee, any use, distribution,
>disclosure or copying of this email is strictly
>prohibited.
>Confidentiality and legal privilege attached to this email
>(including any attachments) are not waived or lost by
>reason of its mistaken delivery to you.
>If you have received this email in error, please delete it
>and notify us immediately by telephone or email. Peter
>MacCallum Cancer Centre provides no guarantee that this
>transmission is free of virus or that it has not been
>intercepted or altered and will not be liable for any delay
>in its receipt.
>========================================================================