Re: Leica resonant scanner-live for cell imaging
Posted by
Rietdorf, Jens on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Leica-resonant-scanner-live-for-cell-imaging-tp590925p590936.html
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Message
Dear Benedikt,
in my opinion,
the major advantage of a parallel compared to single beam scanner is a result
of the low saturation level of the fluorophores (the laser power is split
up to many beamlets or a line or other pattern) and therefore reducing the
population of molecules in dark states, while keeping a high overall photon
flux. In resonant scanners, this beneficial effect (also) results
from avoiding dark states, its descibed in this
article:
Donnert, G., C. Eggeling and
S. W. Hell: |
| Major signal increase in fluorescence
microscopy through dark-state relaxation. |
| Nature Methods
4 81-86 (2007) |
Among
the parallel scanning devices, the spinning disk is (again in my opinion) the
most efficient device, because it avoids turning times of scanners, which are
dead times for the acquisition, because beams will have to be blanked during the
turnaround. Parallel scanners with CCD detectors typically outperform PMTs in terms
of QE over a broad wavelength range.
Point
scanners on the other hand are more flexible, ie you can scan lines or
AOIs, rotate scanfields etc., also confocality is not compromised under any
circumstance or in any direction. For deep-in tissue imaging (above 50microm),
single beam scanners typically outperform parallel scanners.
For
full frame confocal of living specimen of moderate thickness (up to 25microm),
my favorite scanner is still the Yokogawa CSU-22 (no commercial interest) in
combinaton with a EMCCD camera and powerful lasers >25mW.
regards, jens
---
Dr.
Jens Rietdorf[hidden email]
Head
Microscopy
Novartis Research
Foundation
Friedrich-Miescher-Institute, wro1066.2.32
Maulbeerstr.66,
CH-4058 Basel, Switzerland
phone +41(61)69-75172 mobil +41 798284737
Email:rietdorf(at)fmi.ch
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Can anyone compare the performance of
the Leica SP5 resonant scanner with
that of a spinning disk (yokagawa)
confocal system in terms of
- image quality
-
photobleaching/phototxicity
- general "usefulness" for live cell confocal
imaging
The resonant scanner was recently demonstrated to us and I was
quite
impressed. I guess it is not as fast as spinning disk confocals can
be, but
you seem to be able to image at a rate of at least 10 fps at
512x512, which
is all we need. The images we got were very nice, but I
did not have the chance to do extensive tests.
I would appreciate very much comments from anyone who has experience
using the resonant scanner for live cell confocal imaging.
Many
thanks,
Benedikt
Kost
____________________________________________________
Benedikt
Kost, PhD
Warwick HRI
Wellesbourne
Warwickshire, CV35
9EF
UK
Tel: +44 (0)24 7657
5092
Fax: +44 (0)24
7657 4500
e-mail: b.kost@...
internet: http://www2.warwick.ac.uk/fac/sci/whri/