Re: Mercury lamp alignment

Posted by Michael Weber-4 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Stability-issues-with-LAS-AF-1-8-0-tp590963p590968.html

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Dmitry,

you won't be able to see the lamp filaments after passing through the
fiber. So yes, if you want to align the lamp in a proper way, you have
to remove the fiber from the lamphouse and align it at the wall (or an
another stand).

I have to say that I so far did it also based on highest throughput, but
I know about the security issues and will (hopefully) do it better soon.
The fiber is anyway killing a lot of light, so on a longterm view it
might be better to go for other solutions, i.e. LED sources or stronger
Metal Halide lamps with better fibers.

cheers,
Michael


Dmitry Sokolov wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear All,
>
> May I ask you for your suggestion on proper procedure of mercury arc
> bulb alignment of fiber-coupled 106 z lamp housing. I was recommended to
> align it by maximum of intensity of light passing through the objective
> of (Leica DM6000B) microscope . The intensity and uniformity is
> controlled looking on a sheet of paper placed under turret with
> objective removed.
>
> The corresponding "Leica DM6000B DM6000M Operating manual" has a warning
> of lamp explosion by overheating if "The bright tips of the arcs, the
> focal sports..." projected onto each other (p.55) or contact elements of
> lamp. The alignment procedure is based on viewing the arcs through a
> special window of the microscope. In my case I see no images of arcs but
> a number of rings originated probably from the fiber ends.
>
> Should I better remove the fiber coupler and align the lamp looking on
> it's image on the wall?
>
> I would highly appreciate it if you could share your experience.
>
> Cheers,
> Dmitry
>
> ___________________________________________
> Dr. Dmitry Sokolov,
> Manager of Confocal Microscopy Unit
> Institute of Molecular Biosciences
> Riddet Road, Massey University
> Private Bag 11-222, Palmerston North
> Tel: +64(6)356-9099 ext. 5549
> Email: [hidden email] <mailto:[hidden email]>
> Booking: http://confocal-manawatu.pbwiki.com/Booking