The apparent thickness of a layer with a real depth of 'x' is 'x/n', where n is the RI (viewing from vacuum; otherwise n1/n2). I guess the coverslip correction doesn't correct for that; that's why your were surprised, I guess.
Zoltan
On Dec 10, 2007 1:52 PM, S. Pagakis (IIBEAA) <
[hidden email]> wrote:
> From: Confocal Microscopy List [mailto:
[hidden email]]
> On Behalf Of George McNamara
> Sent: Sunday, 9 December 2007 6:08 AM
> To:
[hidden email]> Subject: Re: Coverslip thickness and correction collar ... option Ib
>
> Search the CONFOCAL archive at
>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>
> option Ib. Measure of the glass thickness before you culture your
> cells. Then, for this lens, use dishes whose coverglasses are 170 um
> (or whatever thickness works best for the lens at its current
> "optimal" setting).
>
> If one in five are the right thickness, than the cost of that dish,
> for this experiment, has gone from about $2 to $10. Should take less
> than a minute to measure a dish using a 20x, decent NA, dry lens.
Dear George
On a recent test I did, I discovered something that I had not expected.
A dry objective would significantly underestimate the thickness of a
glass coverslip.
For a #1 coverslip, a 20x dry objective would measure it as 98microns
thick, where as a 63oil objective as 138microns thick.
For a #1.5 coverslip the values were 113microns and 160microns
respectively.
As I said, I wasn't expecting it because the 20 lens was coverslip
corrected.
So you have to be very careful when using dry objectives to measure
anything in Z.
regards
*********************************
Stamatis Pagakis Ph.D.
Biological Imaging Unit
Biomedical Research Foundation, Academy of Athens
[hidden email]
> Compared to the cost of the confocal time (or of an LSM510), or of
> your time to do manual image analysis, or cost of the analysis
> computer and image analysis software, the extra cost is trivial. Plus,
> you can sort the "rejects" and use them with lenses each thickness is
> optimized for.
--
--