Re: triple labeling with antibodies

Posted by Morrison, Ian E on
URL: http://confocal-microscopy-list.275.s1.nabble.com/triple-labeling-with-antibodies-tp591231p591241.html

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Fab fragments are a good idea, but getting them can be very time-consuming and expensive.  Some companies offer this service, but the quality is highly variable - naming no names, we've had some total failures to produce labelled Fab that still binds.  In our own laboratory, producing Fab from IgG varies enormously; MHC class II works well, CD74 seems to be impossible.  You need to make a number of test runs with variable conditions, so shedloads of IgG may be required.

If you have to use a 514nm excitation, can I put in a word for Oregon Green 514 (no commercial interest).  This can label whole antibody up to ~25probes per IgG with no self-quenching, and on the samples we made using MHC antibodies there was no aggregation and the binding was only slightly compromised.

Ian